We tried all the above steps, but the problem continued. We had to change the parameters in the Methods file as suggested in another reply from Cendy D. to make the instrument continue
Fragment analyser (HS NGS kit) fail to achieve 4µA
Hello, anyone here know what happen with my icp-oes?
i got error message '' gas scp write channel" and icp-oes unable to running due plasma ignition failed, the icon plasma also blinked many times.
Thank you.
Rerey
Hello,
Are you sure the end tip of the PLOT column ends up (and 2 mm withdrawn) inside the FID Jet?
Any problems when igniting the FID?
FID Jet could be contaminated from PLOT stationary phase particles ..
Best regards -
Lars
Our lab has an Agilent 1260 Infinity HPLC, and we are encountering an annoying issue with the desktop connected to it. The desktop sometimes shut down automatically, while the HPLC is still running the analysis. The HPLC keeps working and and the analysis…
The vials and caps are the same ones we have been using for years now; they are your standard 2mL GC/LC vials. I opened up LabAdvisor and ran the gripper test. It failed to release the vials even after re-initialization. I removed the gripper caps, wiped…
Hello,
I am new to MassHunter, so this may be a basic question or just something I'm doing wrong. I am trying to build a quantitation method for a large number of compounds we are analyzing via LCMS. I ran our calibrators with our LCMS method and have…
The Pfeiffer turbopump on my 5977C MSD will not start after shutdown; it does not exceed 5%. When I boot the MSD, the green LED on the pump controller is solid. Then when the turbo fails to reach 80%, the LED starts blinking and the MSD panel returns…
Hi
When FID is off means no flame.
Voltage applied during ignition and while FID OFF so ignitor also Off..
When ignitor off then voltage also off...
Hope, it's clear to you..
This Information Applies To: Agilent MassHunter 11 and later
Issue
Generating Acquisition Desktop and Quantitative Analysis shortcuts.
Steps to Follow
Double-click the Control Panel icon on the desktop or select Windows Start menu > Agilent...
This Information Applies To: Agilent MassHunter Workstation for LC/TQ and MassHunter Workstation for LC/TOF and LC/Q-TOF
Issue
How to view the MassHunter Data Acquisition Software Logs entries, change settings, and export the Logbook for archiving...
This Information Applies To: ICP-MS MassHunter with User Access Control (UAC)
Issue
It is necessary to update the database folder path in Change User Access Control:
After creating a new database in SDA with ICP-MS MassHunter.
To view a batch...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Over time, the electrodes may become pitted or eroded and users may see a decrease in precision or begin to experience the "Tube fault" error when new tubes are installed. If...
This Information Applies To: Agilent GTA Furnace and Flame Instruments, Agilent 240FS/240GFAA/GTA120/PSD120
Issue
There may be occasions when users need to remove GTA furnace center block workhead (instruments with both Flame and Furnace capabilities...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
The shroud should be replaced annually, or when there is evidence of damage caused by mechanical shock or contamination from liquid spillage.
Items Required
Shroud Removal...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Before replacing the electrodes and/or shroud, the GTA Furnace workhead will need to be disassembled. After replacing the electrodes and shroud, the GTA Furnace workhead will...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer
Issue
After starting a run, the run is aborted with the following error: Current = 0.0uA, Expected range: 5 - 3000.
Steps to Follow
Check...
This Information Applies To: Agilent OpenLab CDS 2.x in file system configuration
Issue
Sometimes, you will need to change the Projects Root Path / Project Storage Path . This article outlines the procedure on how to change it.
Background
...
This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7
Issue
Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES
Tip : These steps should be completed after the Subsystem Communication...