I am having a similar issue. First, I replaced a needle seat and then the auto-referencing failed because I failed to recognize that the black sample holders needed to be removed. The needle mechanism seemed forever stuck in the parked position. I was…
Hi,
you should set your instrument to "user tune" in startup. But plasma correction is intended for a regualr setup. you can try to run it with your "special configuration".
Under "Startup Configuration" change the Ignition Sequence to "user tune…
Dear colleagues, I keep experiencing the same issue with not being able to open some batches in MH quant (5.1, G7201D / Version D.01.01 / Build 653.5), while others work just fine. I have integrated three batches that were all analyzed, and chromatograms…
Hello, i get unusual high signal on FID on 7890 and this message
"Front detector - Fault high voltage error"
What is happening? There is some part I can try to replace before call a technician?
Thank you
Many GC and GCMS installs do not have the items needed to complete the installation. GC and GCMS system installs rely heavily on site readiness and customers having all the material needed, like copper tubing, Swagelok® fittings, and regulators. This…
Additional Items to add to GC and GCMS Orders-Customer Facing.pdf
I updated to newer version of Masshunter 10 & when I tried to re establish communication in Lab Advisor, It shows the 1260 HPLC components but Could not connect to QQQ . No connection could be made because the target machine actively refused it 192.168…
Hi GeorgiaDean were you able to solve your problem?
In regards to the FID not lighting, be sure to refer to the Troubleshooting an FID not igniting article in the Knowledge Portal.
The struggles continue...
I've been having trouble with my 240fs, and I've arrived at a new problem state. The igniter lights just fine, and the burner head lights, but when I let go of the igniter button the igniter flame goes out immediately and then…
This Information Applies To: Agilent MassHunter 11 and later
Issue
Generating Acquisition Desktop and Quantitative Analysis shortcuts.
Steps to Follow
Double-click the Control Panel icon on the desktop or select Windows Start menu > Agilent...
This Information Applies To: Agilent MassHunter Workstation for LC/TQ and MassHunter Workstation for LC/TOF and LC/Q-TOF
Issue
How to view the MassHunter Data Acquisition Software Logs entries, change settings, and export the Logbook for archiving...
This Information Applies To: ICP-MS MassHunter with User Access Control (UAC)
Issue
It is necessary to update the database folder path in Change User Access Control:
After creating a new database in SDA with ICP-MS MassHunter.
To view a batch...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Over time, the electrodes may become pitted or eroded and users may see a decrease in precision or begin to experience the "Tube fault" error when new tubes are installed. If...
This Information Applies To: Agilent GTA Furnace and Flame Instruments, Agilent 240FS/240GFAA/GTA120/PSD120
Issue
There may be occasions when users need to remove GTA furnace center block workhead (instruments with both Flame and Furnace capabilities...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
The shroud should be replaced annually, or when there is evidence of damage caused by mechanical shock or contamination from liquid spillage.
Items Required
Shroud Removal...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Before replacing the electrodes and/or shroud, the GTA Furnace workhead will need to be disassembled. After replacing the electrodes and shroud, the GTA Furnace workhead will...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer
Issue
After starting a run, the run is aborted with the following error: Current = 0.0uA, Expected range: 5 - 3000.
Steps to Follow
Check...
This Information Applies To: Agilent OpenLab CDS 2.x in file system configuration
Issue
Sometimes, you will need to change the Projects Root Path / Project Storage Path . This article outlines the procedure on how to change it.
Background
...
This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7
Issue
Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES
Tip : These steps should be completed after the Subsystem Communication...