I have the air generator and his out pressure consist 52 psi, and and air consumption is 1000 ml/min. the EPC module GC 7890b work in rang 55-80psi. I have 2 detector FID. Tell my - with this generatorom may be worked ?…
yes, it's the purged 2-way splitter. Bottom is column entry, then detector 1 is FID, detector 2 is qtof. (switching FID and qtof position doesn't change anything)
We had 3.7mx0.15mm to qtof and 1mx0.15mm to FID with constant flow 0.8 ml/min for a 1…
Hi all,
I'm using a 0.18 mm HP-5 GC Column on a Shimadzu GC-2014 connected with a FID. I'm looking to get new ferrules (the old ones had been passed down from previous users) and wanted to discuss some options. I tried using some 0.5 mm graphite ferrules…
Thanks Marty, I was afraid of that.
Too bad really since piecewise is the most conventional kind of multi-level calibration.
I train using the example of a typical pH meter where they still naturally calibrate at pH 4, 7, & 10. Therefore there…
You can swap out cylinders to do it, but that means a lot more tubing, etc. to purge when you want to run. With the EPC unit, you would simply load the appropriate method to use it and the switch is done at the instrument. Much less tubing to purge and…
Hello,
To calculate the RSD and Pass/Fail results, one can try the following:
In the reporting go to the edtor and open "table properties"
In my specific case, I want to calculate RSD for Area, so I choose the column A, pA*s and press "Column…
Hi everyone, My MP-AES 4210 Machine sa is failing to suck/take-up the samples from the sample tubes..
I tried changing the sample pipes but nothing worked. Kindly asking for help on how to sort out this problem?
Regards,
Kalunga
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...