Is is possible to create a in built library in GC7820A with FID Detector using Chemstation Edition?
Is is possible to create a in built library in GC7820A with FID Detector using Chemstation Edition?
Hi James
Sorry I should have not use the word library. It is only for the internal purpose just to identify the peak and it is not for the quantification and it is only for the quality check purpose.
Thanks
Revathy
Hi revathy,
Libraries are usually for three dimensional data such as MSD data where you can extract a spectra. In the case of two dimensional data from a FID, the best you could do is to create a calibration table from known standards and use that for identification by retention time and concentration. This is not a library , but simple identification on your system using standards . I hope this helps.
Regards
James
Hi Revathy,
I have marked this as assumed answered due to its age. Feel free to post again with questions if needed.
Regards
James
This question has been marked as assumed answered.