Expression for calulating RSD in OpenLab 2.7

I'm looking for a couple expressions in the reporting module where it would take the peak area of a certain compound for all the selected injections included iin the report, and then calculate RSD.  In another field, it would compare that RSD against a set value and determine if it was a pass or fail.

Parents
  • Hello,

    To calculate the RSD and Pass/Fail results, one can try the following:

    In the reporting go to the edtor and open "table properties"

    In my specific case, I want to calculate RSD for Area, so I choose the column  A, pA*s and press "Column Properties"

    In the "Summary Calculations" select "Average", Standard Deviation" and "RSD" checkboxes. For all of this values one can select how many digits after comma it will be rounded.

    In this screen I would also create a "variable" for the RSD with the name "arearsd" you can use your own just remember it.  We we will use it later for "Pass/Fail". After this press OK and OK.

    Press this button.

    And go to Report parameters, and create new propertie by pressing "Add" button and filling Parameter name, Prompt name and Value.

    Parameter name will be used for "Pass/Fail". Value: you can choose later, just put default value. Prompt name is how this propertie will be displayed. After Press OK

    Now from Report Items Drag and Drop Text box from left to your report

    In the First box Type Pass/Fail while in another press right button on the mouse and select "Properties".

    Choose "Value" and "fx" to open Expression Editor".

    Here type the following:

    =IIF(arearsd<RSDarea, "Pass","Fail")

    and press OK.

    Refresh your template. Now you should have this box in the top of your screen. You can change this value here.

    Below you can see the Pass/Fail result and Average, SD and RSD values.

    If I change RSD area to 0.8 I will get Fail result

    Important that report type must be Single sequence summary (data grouped per sequence)

    Below is my version of SW I used

    Hope this was helpful.

    Kind regards,
    Rustam

Reply
  • Hello,

    To calculate the RSD and Pass/Fail results, one can try the following:

    In the reporting go to the edtor and open "table properties"

    In my specific case, I want to calculate RSD for Area, so I choose the column  A, pA*s and press "Column Properties"

    In the "Summary Calculations" select "Average", Standard Deviation" and "RSD" checkboxes. For all of this values one can select how many digits after comma it will be rounded.

    In this screen I would also create a "variable" for the RSD with the name "arearsd" you can use your own just remember it.  We we will use it later for "Pass/Fail". After this press OK and OK.

    Press this button.

    And go to Report parameters, and create new propertie by pressing "Add" button and filling Parameter name, Prompt name and Value.

    Parameter name will be used for "Pass/Fail". Value: you can choose later, just put default value. Prompt name is how this propertie will be displayed. After Press OK

    Now from Report Items Drag and Drop Text box from left to your report

    In the First box Type Pass/Fail while in another press right button on the mouse and select "Properties".

    Choose "Value" and "fx" to open Expression Editor".

    Here type the following:

    =IIF(arearsd<RSDarea, "Pass","Fail")

    and press OK.

    Refresh your template. Now you should have this box in the top of your screen. You can change this value here.

    Below you can see the Pass/Fail result and Average, SD and RSD values.

    If I change RSD area to 0.8 I will get Fail result

    Important that report type must be Single sequence summary (data grouped per sequence)

    Below is my version of SW I used

    Hope this was helpful.

    Kind regards,
    Rustam