Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Realize this was a while ago but I just came across it; wrote something similar a while back that processes the data in a given sequence folder - could easily be adapted to run as post-run and post-sequence or also with a specified folder. Hopefully this…
Thank you Paul! I haven't got a chance to try it yet. But I found the sample might carry too much salt (in heptane) that may damage the column and system. The drop always occurs right after one of the std peak which lead me to think it is somehow LRI…
DP-3 Dry Scroll Pump puts distracting noises, oily messes, and pump seizures behind you once and for all. This is the second of a series of instructional videos for the IDP-3 Dry Scroll Pump.
For more information, please go to www.agilent.com/chem/idp3…
Hi, thank you for your patience! I hope the following instructions help you with your question; not sure what exactly is meant by "processing time". Unfortunately, I cannot see your attached picture.
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...