This information applies to: Agilent 5977B, 5977C GC/MSD and 7000D, 7000E, 7010B and 7010C GC/TQ
Issue
Through the front panel instrument status LED, the operator can check the status of the instrument using a combination of color codes and LED...
This information applies to: Agilent GC/MS systems running MassHunter data acquisition software for GC/MS (B.07.x - B.10.x).
Issue
Set up a GC only (without the mass spectrometer) configuration in MassHunter.
Steps to Follow
Close any acquisition...
This Information Applies To: Agilent GC/MSD systems
Issue
Instructions on how to vent the Single Quad GC/MSD instrument using OpenLab CDS software.
Background
If the GC/MSD is running and under vacuum, then the analyzer needs to be vented...
This Information Applies To: Agilent GC/MSD Systems using the Agilent IDP-3 dry scroll pump
Issue
Scroll pump requires replacement two tip seals and one O-ring after approximately one year of use (9000 hours).
Requirements
The system must...
This Information Applies To: 5977A and 5977B
Content: This procedure is to guide the installation of a capillary column on 5977A/B GC/MS with a standard column nut or self-tightening nut.
Warning : a. The GC and MS operate at high temperatures...