This Information Applies To: Agilent GC/MSD systems
Issue
Instructions on how to vent the Single Quad GC/MSD instrument using OpenLab CDS software.
Background
If the GC/MSD is running and under vacuum, then the analyzer needs to be vented...
This Information Applies To: Agilent GC/MSD Systems using the Agilent IDP-3 dry scroll pump
Issue
Scroll pump requires replacement two tip seals and one O-ring after approximately one year of use (9000 hours).
Requirements
The system must...
This Information Applies To: 5977A and 5977B
Content: This procedure is to guide the installation of a capillary column on 5977A/B GC/MS with a standard column nut or self-tightening nut.
Warning : a. The GC and MS operate at high temperatures...
This Information Applies To: Agilent 7000A/B/C/D Triple Quadrupole GC/MS systems
Issue
When a triple quadrupole GC/MS system tune fails, the user can check for any abnormalities by setting the tune file to default tune file before checking for hardware...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...