Dear, Sr
When I want to go to execute any program in WinUV software, it gives a message "Error code 9:1 CSM error,CSM error"
It seems a comunnication error, the instrument is connected via GPIB-USB-HS.
Anyone knows how to do?
I assume it is packed tight with insulation? Otherwise I'd just call Agilent for any ideas, maybe a board is bad or something and you need a service call.
Is there going to be 60 day demos for these products? The previous versions are so buggy that taking a leap of faith and spending more money on these products hoping that there bugs are fixed isn't something that one should be doing. Kevin
Hi. As Chris says, if the flow is 12 ml/min means you have a massive leak.
But, just in case, did you close the Helium cylinder, right? That's a part of the procedure.
Also, I think that 2 years without cleaning the lenses is quite a lot of time…
Hi all,
I was just wondering whether anyone has experienced issues when reducing MS Scan time down from the 500ms default. I am running MS2 scans for method development however whenever I decrease the scan time, the work list does not progress and the…
Hello team
I write this message in order to receive some help regarding the use creation of reports using the Mcrosoft Add-in of a customer's Microlab 5.7. It is supposed that these icons must appear in Microsoft Word in order to create reports but…
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...