A 5% H2 blend should work also. If you're using multiple calibration standards, make sure the range of the standards matches or exceeds the expected range in your samples.
One thing to watch for: I have seen instances where H2 saturates the TCD at high…
Offsetting the axes, Y or X, is not a feature in MassHunter. You can do head to tail comparison.
Overlaying in different colors and choosing which one is active can be done but there's not a way to do what you're asking even in a custom report.…
Good morning. We analyze fatty acids and The compound response factor of each fatty acid before the softwere changed remained fixed now it changes. I tried to copy the various parameters set in the other gas chromatograph which has its own softwere, but…
Hello,
I'm having the same problem with the overwhelming amount of orange highlighting for samples where the qualifier peak was not found. I tried the suggestion above of enabling "Ignore Peaks Not Found" in the method and re-applying it to the batch…
The tune evaluation was written for an orthogonal ion source (SS/Inert/Extractor) system, atune or etune, running helium carrier gas between 0.8 and 1.4 ml/min, at 230° C ion source and 150° C quad temperatures. It may not pass using other temperatures…
Hello,
I am not sure which Agilent software (CDS 2.x , Chemstation, EZChrom) you are using, but most customers use third party software for simulated distillation and other RT vs BP petrochemical calculations. If your need is simpler, you could do…
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...