Hello OliwiaK ,
I apologize for the delayed response.
Could you please let us know what program you are using, such as Qual or Quant, and what library you are trying to utilize.
Problem solved. The format in Region settings should be set to "English (United States)", nothing else. As a result, the Decimal symbol under Numbers in Additional settings becomes a point (Anglo-Saxon decimalization) instead of a comma (Western European…
Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG
Issue:
Only plasticware from the approved consumables list should be used in the Agilent Parallel Capillary Electrophoresis systems. The use of plasticware...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Question:
Can more than one Agilent 2100 Bioanalyzer instrument be connected to one PC?
Steps to follow:
It is possible to connect a maximum of two Bioanalyzers to one laptop...
This Information Applies To: 2100 Bioanalyzer, Bioanalyzer Chip Priming Station: part number 5065–4401
Maintenance procedures and how frequently they are required :
Procedure
Time interval
Or if this issue occurs
Where to find a detailed...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue:
Running the hardware diagnostics on the Agilent 2100 Bioanalyzer system can help to investigate whether the instrument performs as expected.
Steps to follow:
Required...
This Information Applies To: 2100 Bioanalyzer, RNA 6000 Nano kit, RNA 6000 Pico kit, Small RNA kit
Issue:
Is it possible to size RNA on the Agilent 2100 Bioanalyzer system?
Resolution:
Agilent offers total and small RNA assays on the Bioanalyzer...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue:
The Agilent 2100 Bioanalyzer system does not detect a prepared microfluidic chip after it is inserted into the instrument. The interface shows the instrument...
This Information Applies To: 2100 Bioanalyzer, Bioanalyzer chips
Issue:
Can 2100 Bioanalyzer assay chips and reagents be interchanged?
Resolution:
Some assay chips can be used with multiple assays, however some cannot because the chip layout...
This Information Applies To: Agilent MassHunter GC/MS Acquisition software
Issue
The MassHunter software performs several checks during the startup sequence. If it detects a problem, an alert message box will be displayed and the message bar on...
This Information Applies To: Agilent MassHunter Quantitative Data Analysis Software
Issue:
The "NO DATA POINTS" error message appears in the Compound Information window in the lower-left corner in Quantitative Analysis. Sometimes, this problem may...
This Information Applies To: Agilent GC/MSD MassHunter with MSD ChemStation Data Analysis
Issue
When you are using the Data Analysis Report, you may need to specify a certain number of decimal places for your results different to the default decimal...