Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Articles Agilent Parallel Capillary Electrophoresis Compatible Plasticware
  • Articles
  • Files
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Automated Electrophoresis Portal Home
  • Automated Electrophoresis Portal Articles
  • Automated Electrophoresis Portal Files
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Automated Electrophoresis Systems
  • Component :  Femto Pulse Systems Fragment Analyzer Systems Oligo Pro II Systems ZAG DNA Analyzer Systems
  • Product Name :  5200 Fragment Analyzer System 5300 Fragment Analyzer System 5400 Fragment Analyzer System Femto Pulse System ZAG DNA Analyzer system
  • Task :  Operation
Related Articles
  • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, Oligo Pro II Issue: This article explains the steps to take after installing a brand new array in the Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, and Oligo Pro II systems. Steps to follow Capillary array replacement steps are detailed in chapter 8 of the Fragment Analyzer user manual , Femto Pulse user manual , ZAG DNA analyzer user manual , and chapter 9 of the Oligo Pro II user manual...
  • Insert Disk Error in the Agilent Fragment Analyzer Controller Software

    This Information Applies To: 5200/5300/5400 Fragment Analyzer controller software, Femto Pulse controller software, ZAG DNA Analyzer controller software, Oligo Pro II controller software Issue: The Agilent Fragment Analyzer, Femto Pulse, ZAG, or Oligo Pro II controller software does not open when the desktop icon is double-clicked. Instead, one of the following errors come up: "Insert the 'Fragment Analyzer' disk and click OK" "The feature you are trying to use is on a network resource...
  • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, ZAG, and their respective controller software Issue: A full power cycle can help to recover communication/connection issues of Agilent parallel capillary electrophoresis instruments (5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG systems). Communication and connection issues can materialize already during software start as connection failed error messages although the instrument is...
  • Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software. Issue Failed to build pressure error comes up shortly after starting a separation. Resolution Failure to build pressure error occurs when the system is trying to build pressure when liquid is being pumped during conditioning or flushing of the instrument. Possible causes and suggestions to address the issue are: 1. Empty reagent...
  • Run Aborted and Target Z Position Error on Parallel Capillary Instruments

    This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Agilent Femto Pulse, Agilent ZAG DNA Analyzer Issue: The instrument encounters a run error, Run Aborted, Position fault: Did not reach target Z position ( Figure 1 ) . This error can occur if the drawer gets opened when the stage is still moving. The error occurs soon after starting the run. Figure 1. Run Error:CtrlTaskStage Error Resolution: System Reboot: Perform an entire system reboot as described in the...

Agilent Parallel Capillary Electrophoresis Compatible Plasticware

Answer

This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG


Issue:

Only plasticware from the approved consumables list should be used in the Agilent Parallel Capillary Electrophoresis systems. The use of plasticware with different dimensions could lead to decreased injection quality and consistency as well as damage to the tips of the instrument capillaries.


Resolution: 

Compatible plates and tubes are listed in the following manuals:

  • 5200, 5300, and 5400 Fragment Analyzer System Manual page 141-144
  • Femto Pulse User Manual page 141-142
  • Oligo Pro II User Manual page 120-121
  • ZAG DNA Analyzer System Manual page 140-141
 Tip 
Learn how to effectively operate your Agilent parallel capillary electrophoresis system:
5200, 5300, and 5400 Fragment Analyzer System Manual
Femto Pulse User Manual
Oligo Pro II User Manual 
ZAG DNA Analyzer System Manual 

For Research Use Only. Not for use in diagnostic procedures.

PR7000-8084
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help