This Information Applies To: Agilent MassHunter GC/MS Acquisition software
Issue
The MassHunter software performs several checks during the startup sequence. If it detects a problem, an alert message box will be displayed and the message bar on the bottom left of the software window sometimes shows the valuable clue for error.
Resolution
The Alert dialog box appears with the message: "The ChemStation did not complete its initialization cycle". Click OK to display the instrument log file.
The entry in the log file indicates the source of the error. The common sources of the error that are listed in the error log include:
Not all the cases and their solutions are included in this article. Contact your local Agilent Online Technical Support for phone support for further possible solutions.
Case 1: MSD: Communication Error: caused by broken filament
Error in log file: MSD: Communication Error: Download command timed out ('SLICK_BEGIN[192.168.254.10] ABCDEF12'). Error connecting to MSD from Mscfg - SCIL Download command timed out (Figure 1).
Figure 1: Log message indicating MSD communications error
Power cycle the MS and restart the software.
If the error recurs, open the instrument Current Logbook mslogbk.htm file under the path x:\Masshunter\GCMS\n\
(x is the installation drive of MassHunter folder, n is the instrument number).
If the error: MSD Fault - 0X0000000000000008 No Emission Current is displayed, the filament is no longer emitting current and must be replaced (Figure 2).
Figure 2. Current Logbook message after power cycling indicating the filament needs to be replaced.
Solution: Change the filament.
Case 2: Invalid Sequence file name or Sequence file corrupted
Error in log file: D:\MassHunter\GCMS\1\sequence\2019.sequence.xml invalid--couldn't open the file 'D:\MassHunter\GCMS\1\sequence\2019.sequence.xml'.
Solution:
Tips:
|
Case 3: Incorrect account permissions (typically corporate IT changed the computer and account permissions).
Error in log file: 1 Access to the path 'D:\2020\GCMS\X2P-5695\Method\ALS-5695-JYD-624-SIM-bf1-1.M\method.xml' is denied (Figure 3)
Figure 3. The error message in the log file
Solution: Switch to the Manager account with full read/write/execute permissions on the local PC and restart the software.
Case 4: The MassHunter GC/MS Acquisition software is configured with a third-party sampler
Solution:
|
Learn how to effectively operate your Agilent MassHunter software:
Agilent MassHunter GC/MS Data Acquisition for the 5977 Series MSD System user manual on Agilent.com.
|