Hi hennindm, Congrats on your 8890 purchase! The new GCs are designed to be used both in local settings with a computer near the instrument, and remotely, so the best fit was to build the data access into a workflow that permitted both, aka the browser…
Hello, I updated the tags on this post to increase visibility. The first thing I would recommend would be to run the Software Verification Tool (might be called the Installation Qualification Tool or IQT in your revision) found by going to Start>All Programs…
Martin, I have run into one more problem. In order to display the compound as a percentage, I also need to get the average of another field, but this one is a custom field( "Sample Weight/g") . I tried using the same query that you created to get the…
Hello, If you want to approach this from an intelligent report angle, then you can set up a Pass/Fail flag on a Sequence Summary Report using the IIf() function. The tricky part would be to make sure that the logic in the Pass/Fail expression is referencing…
Hey guys, Our gradient test has been failing, got a new MCGV, replaced inlet and outlet filters, rotor seals, frits, switched degassers, switched lamp, cleaned flow cell, and still get a lot of drift...particularly when more than one channel is used.…
Hey guys, I'm calibrating a UV/Vis, instrument model 8453, and I seem to consistently give varying results for the photometric accuracy test (with potassium dichromate). The protocol requires it be done under 4 minutes. My question is, why? Is it unstable…
You can program a gradient to go to 100%B in theory. The BPR control will not be great at 100% co-solvent but its possible. However, I have not seen an application which requires more than 50% MeOH, EtOH or IPA co-solvent. Maybe others on here have specific…
Hi Kristen, Thanks for the speedy reply and links, really useful. Everything parts wise appear OK. The ALS syringe (5ul gold syringe barrel ( 5182 0836) with 0.32um 26-32/42 cone needle (5182-0831)) is listed under 7673/7683 On Column autosampler syringes…
Hello, I recently replaced a SmartCard in my 7700x and have not been able to communicate with the mass spec via MH. The IP settings appear to be correct for the peer to peer connection (192.168.1.127 PC ; 192.168.1.128 7700x ; 255.255.255.0 subnet masks…
This Information Applies To: Agilent GC Systems
Issue
This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler...
This Information Applies To: Agilent GC systems
Issue
Peak tailing is present in a chromatography peak when it has an excessive asymmetry with a trailing edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector)
Issue
Problems observed with thermal conductivity detector (TCD) results can have different underlying causes.
Resolution...
This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000
Issue:
This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out.
Cause:
If the FID...
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue:
The Flame Ionization Detector (FID) jet could become clogged with ferrule material, column bleed, or heavy molecular analytes. The clogged jet will affect...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different high concentration, split analysis techniques including:
General use liners
Auto inject, general use
Auto...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different splitless trace analysis techniques including:
General use liners
Pressure, temperature, volume (PTV)
Hot...
This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet
Issue
Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet...
This Information Applies To: Agilent GC Systems with Micro-ECD.
Issue
The Wipe Test is an annual test that performed by end user of the micro-ECD (Micro Electron Capture Detector or uECD) to help ensure radioactive safety when using the detector...
This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD)
Issue
Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.