This may be due to several factors. First of all, make sure that all plasma monitor connections and cables are properly connected and not damaged. Also check that there is sufficient gas flow (eg argon) to maintain the plasma. Check the cleanliness and…
Hello all,
I´m facing a Silicium contamination in ICP-MS analysis, I have used a 7800 ICP-MS.
I have digested my samples with HF, HCL and HNO3 and neutralized it with boric acid, after the analysis I have Silicium CPS has increased around 30%.
…
Hi Layden,
your source has died (normal ratio Amps/Volts should be starting at 1.9/2.7). Each year you will find a decline in Amps until it has reached 1.6, then it will give a red bar. In your case it's 0, so the filament is probably broken - your…
I have almost the same system as shown in the picture for over a year now. The application was supplied by Da Vinci Laboratory Solutions in the Netherlands. It is a Mettler TGA 2, with a Gerstel MPS autosampler on a Agilent GCMS(MS) with a CIS injector…
Thank you. I contacted the support. I just had to login into agilent subscribenet and download the latest version after entering the authorization key.
Hello,
That looks like a standard IS calibration using area percent. The purity of the standard can be entered as the compound multiplier. I would start by searching the help for calibration, internal standard, and mass percent calculations. You can…
Hello Lucas_S ,
Based on your follow-up explanation in this discussion -
(+) Macro for Batch Export of CSV in OpenLab CDS C.01.10 to a unique archive. - Forum - Chromatography Software - Agilent Community
I don't believe this macro could be easily…
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue:
The Flame Ionization Detector (FID) jet could become clogged with ferrule material, column bleed, or heavy molecular analytes. The clogged jet will affect...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different high concentration, split analysis techniques including:
General use liners
Auto inject, general use
Auto...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different splitless trace analysis techniques including:
General use liners
Pressure, temperature, volume (PTV)
Hot...
This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet
Issue
Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet...
This Information Applies To: Agilent GC Systems with Micro-ECD.
Issue
The Wipe Test is an annual test that performed by end user of the micro-ECD (Micro Electron Capture Detector or uECD) to help ensure radioactive safety when using the detector...
This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD)
Issue
Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items...
This Information Applies To: Agilent GC Systems with flame ionization detector (FID)
Issue
Baking out the flame ionization detector at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items required...
This Information Applies To: Micro Electron Capture Detector (uECD)
Issue
Baking the micro electron capture detector (µECD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Warning ...
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue
This article will describe how an operator can configure the makeup gas on both the instrument keypad and the software driver (Method Editor).
Background...
This Information Applies To: Agilent HPLC systems
Issue
The presence of ghost peaks can be a symptom of a system contamination issue, caused by accumulation of sample and/or foreign substances in the flow path. It is critical for high sensitivity...