Yeah, that's what I thought as well but the AUX EPC flow controllers only are capable of pressure regulation, not flow correct? The problem we were having was since the outlet is connected to the collision cell under vacuum, and since the n2 flow is so…
Hi everyone,
Our lab own a microGC 990 with a PORAPLOT U column. The software we use is Openlab.
I usually run my samples with "sequence" mode. Is there a way to put a "delay time" between samples? I just want a delay between finishing sample 1 and…
In the Offline Worklist Editor for MH12, when using the "Edit Samples" window to input the sample positions, the "Current Configuration" box in the top left isn't shaded out, so that I can't choose it. When using Acquisition to do this same function,…
5 psi inlet pressure running hydrogen with your configuration leads to 2.3 ml/min to the MS and 1.02 ml/min to the FID.
You are welcome. Are you using a two-hole ferrule in the injection port or a splitter of some type? The two-holed ferrules are..…
Which liner is being used? Check out: GC Liner Selector | Agilent You should think about trying one with a frit instead of glass wool, like this one: 5190-5105 | Agilent Or a single tapered, ultra inert liner 5190-2295 | Agilent .
Splt/Splitless Inlet…
Hi JPhillips - in our latest software (QTOF Acq 12.1) - it is much easier to copy and paste the entire "Sample Name" column into the data file column. All rows in the data file column will automatically receive the .d suffix in one action (the paste command…
Hello GUX ,
For these types of errors the first thing to try is resetting the user preferences for Profinder. To do this, first close all copies of Profinder and then navigate to the folder
C:\Users\XXX\AppData\Roaming\Agilent_Technologies,_Inc\ …
The error means that the maximum stroke length of your pump drive is being shorter than it should be.
This can be either due to mechanical issues like a broken piston or an offset in the pump drive gear leading to a physical block before the normal…
Hello aafzal3 ,
I apologize for the delayed response.
You can use either find by method to attempt to find your targets, but to troubleshoot the issue you need to find your target manually by extracting the EIC for the expected molecular ion. Once…
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power on the Agilent ICP-MS Foreline Pump when fully powering on the Agilent ICP-MS System, or after performing maintenance.
Background
The ICP...
This Information Applies To: Agilent ICP-MS MassHunter Software version 5.1 and above
Issue
The user manuals for Agilent ICP-MS Systems for MassHunter 5.1 and above are included in the ICP-MS MassHunter Help and Learning Center. There are manuals...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Elevated oxide ratio values are defined as greater than 1.5% for x-lens configuration or greater than 3% for s-lens configuration. Possible causes are contamination...
This Information Applies To: Agilent GC/MS Systems
Issue
The software version is a valuable information usually required for troubleshooting, software updates preparation or to check installed options.
Table of contents
MassHunter GC/MS Acquisition...
This Information Applies To: Agilent LC modules with RFID tag readers
Issue
Agilent uses RFID (radio frequency identification tag) tag readers in some HPLC components such as detector lamps and flow cells, pump heads, purge valve and mixers. However...
This Information Applies To: Agilent LC systems with Lab Advisor software.
Issue
In Lab Advisor, the Logs and Results register is available for gathering history and additional information on the system. It presents data collected from the configured...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
What information does the ICP-MS instrument status indicator provide?
Background
The ICP-MS Instrument Status Indicator on the instrument, and in the...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power off Agilent ICP-MS systems before maintenance, for long-term shutdown, or for power cycling the instrument. Power cycling may be required when...
This Information Applies To: Agilent OpenLab CDS
Issue
Steps to close an instrument connection to a primary instrument controller, to start a successful connection to a secondary instrument controller, or to configure the instrument.
Background...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or replacement of consumable...