I guess checking the settings will be the first thing. On the GC side, since RT did not shift too much, I think oven temp would be of a less issue. But flow and pressure might act more in this case. Based on info I learnt from here, both pressure and…
Hello AJohnson13 ,
Lab Advisor can be currently downloaded from this page.
Lab Advisor Software Download | Agilent
This may change over time. You can search the web for 'agilent lab advisor download' to find a link to the current download location…
Hello,
So, you should be able to add another column to your report table and change the value to = Compound_Concentration * Sample_Amount / 100 to remove the sample amount and percentage. If you need to remove a multiplier or dilution you will need…
Hello,
LIMS ID is not available as a token for sample name, datafile name, or in CDS DA for report output. You can see the options in the screenshot below. It can be a value in the CSV file but not used in the name. The only other option would be…
Our G4220A (1290 Bin Pump) is setup to work with the seal wash currently. It operates with our instrument with no issue. If I go into the instrument configuration and check the box to enable the pump seal wash (The pump is installed), it will lose connection…
Hi RDP0314,
One thing to check is the injection settings. This behavior is common when the GC is expecting a series of manual injections. To check, click on the Instrument menu shown on your screen capture above and 'Select Injection Source' or similar…
This Information Applies To: 4150 TapeStation, 4200 TapeStation
Issue
How to deal with a TapeStation Electrode Cartridge change is due warning message.
Background
The Agilent TapeStation controller software will notify users to replace the...
This Information Applies To: Agilent OpenLab CDS
Issue
Sometimes the integration result is not ideal for all injections. This article outlines the step-by-step instruction to manually integrate peaks.
Background
There are 2 ways to integrate...
This Information Applies To: Agilent OpenLab CDS , Agilent OpenLab ChemStation
Issue
Calculation of Area percent or Height percent for uncalibrated compounds in OpenLab software. If all components respond equally in the detector, then Area% and...
This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900)
Issue
How to power off the Agilent ICP-MS Foreline Pump to perform maintenance, or when fully powering off the Agilent ICP-MS System.
Background
The ICP-MS MassHunter...
This Information Applies To: Agilent ICP-MS (7500,7700,8800)
Issue
How to check the E2M18 or E2M18F foreline pump to determine if the pump needs maintenance?
Background
The oil cleanliness and oil level in the E2M18 or E2M18F foreline pump...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to replace the foreline pump oil in the DS 402.
Background
The DS 402 oil should be replaced at least every 3 months or when the oil color indicates it needs to be changed...
This Information Applies To: Agilent 8890/8860 GC Systems
Issue
How to delete data that is stored on the GC instrument.
Steps to follow
Caution : Before deleting data from your GC, Agilent recommends extraction of data as a backup...
This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, Femto Pulse and ZAG DNA analyzer
Issue
When no data is available in the Agilent ProSize data analysis software for one or more capillaries, the capillary(ies) could be misaligned...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to check the MS40+ or MS40-S foreline pump to determine if the pump needs maintenance?
Background
The oil's cleanliness and level, in the MS40+ or MS40-S foreline...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Powering on Agilent ICP-MS Systems after maintenance, long-term shutdown, or for power cycling.
Steps to Follow
Confirm that the ICP-MS power plug...