Hello,
Having sorted out our printer problem with MassHunter for GC we started having another issue on Friday night last week. The software suddenly stopped mid run giving an error that the drivers for the MSD could not be found. After restarting…
Hi,
Recently Run the Wavelength accuracy and signal to noise at FLD detector for qualification purposes but it not passing. In wavelength accuracy test got Excitation value 350 it ok but in case of emission got 393 nm its failed required minimum 394…
Hello mr.white, You can do this with a sequence summary report template: Use 'Multi Signal Plot Overlaid' as you've done already; in the properties, the signals need to be overlaid. You may group by signal e.g. FID Front signal/ FID Back signal or, if…
Hi. The pump made loud noises for several days. After that, the pump stopped working normally. Near 600 rpm the pump stops and A-PLUS Software was showing fail status (205) with error code 206 (bit 128) - too hight load. We use Agilent pump - TPS compact…
Thanks for advising, do other errors show up in the logbook?
Please watch the movement of the sampler arm during initialisation:
- Is it shaky or jittery? If yes, turn off module and clean transport rods using IPA & lint free cloth.
- Which position…
Hello,
I will assume you mean you are using the backup restore tool for CDS 2.4 not the ECM XT import scheduler. If you have setup an import scheduler task to back the ECM-XT system up that would be self-referential and would not actually back anything…
Hello,
We have been having issues with our ICP MS performance and tune reports. Both have been failing with low counts. We have changed the peri-pump tubing, remade all standards, remade the tune solution, replaced the torch, and cleaned the cones…
Hi dubang69, Im also using export to CDF in OpenLAB 2.3 and OpenLAB 2.2 (with Data Export Addon). Unofrtunatelly second step for these revisions is to open CDF in some open source software as noreuter recommends. Im using lightweight utility UniChrom…
Hello community, I set up an Agilent 6890 (Software: ChemStation A08.03). A few seconds after I start the GC, the error message "Shutdown ( # 6 ): Back Inlet Flow Shutdown" is shown on the display. The settings of the Back Inlet shows that the pressure…
Hello, I have an issue when I try to install OpenLab CDS EZChrom Version A.04.07 The plan is to install EZChrom on a Citrix to allow the user to retrieve files from ECM for review or reintegration. I start the installation and select the Installation…
This Information Applies To: 5975 and 5977 GC/MS systems with MassHunter Acquisition software
Issue:
High-quality GC/MS data requires a low, and stable background of air and water present in the MS analyzer.
This article demonstrates a systematic...
This Information Applies To: Agilent 5975 and 5977 GC/MSD systems with MassHunter Acquisition software
Issue:
High-quality GC/MS data requires a low, and stable background of air and water present in the MS analyzer. This article describes how to...
This Information Applies To: Agilent 5977A/B GC/MSD with MassHunter GC/MS Acquisition Software
Issue:
This article describes the procedure to safely vent (shut down) the 5977A/B GC/MSD with Electron Ionization (EI)
The MSD will be damaged by incorrect...
This Information Applies To: HES ion source in 5977B/C GC/MSD, 7010 triple quadrupole GC/MS systems
Issue
The filaments are replaced as part of routine maintenance, or when they break (zero output current).
Items Required
Gloves, nylon,...
This Information Applies To: Agilent 5977A/B/C GC/MSD
Issue
The Electron Multiplier Horn must be replaced when it is contaminated or the EMV value reaches 3000 V with a clean ion source.
Items Required
Electron multiplier horn (5977B/C:...
This Information Applies To: Agilent 5973/ 5975/5977 GC/MSD with rotary vane foreline pump
Issue:
The oil mist filter is recommended to be replaced annually. If the color of the filter has changed or there is smell from the filter, the oil mist...
This Information Applies To: Agilent EI Source
Issue:
During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments...
This Information Applies To: Agilent G7129A/B/C Vialsampler
Issue:
Replacing the needle and needle seat of the Vialsampler is required when:
The needle is visibly damaged.
Leaks or blockages in the vialsampler are observed, especially between...
This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps
Issue:
How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps.
...
This Information Applies To: Agilent UV and RID detectors
Issue
A clogged flow cell causes high backpressure. The clog can be removed by flushing the system with solvents at high pressure. If you suspect a clog or experience high backpressure due...