Hello,
I am getting the following error on OpenLab CDS Chemstation Edition (Rev. C01.10 [287]): "User Change (Unlock) failed, because instrument access is not permitted for user. Logging Out User". We are initially able to login once, but then when…
Hello mdeighan ,
Manual integration is limited to the rules of the selected integrator and the settings of your quant method. The error and behavior you are seeing typically indicate that the peak you are attempting to create is too far outside of the…
The mass calibration/check for the quadrupole is failing across all isolation widths. Mass delta on 322m/z exceeds limits.
We system tuned a few weeks ago. Can someone please outline the troubleshooting workflow for this type of issue (where to start…
I want to check Total hydrocarbon by FID without identifying the hydrocarbon species . Agilent suggest to give me DB-1 15m, 0.32mm, 0.10u as a non polar column to elute all HC without separation as one peak . Does this column will work for Total hydrocarbon…
If you receive an onscreen notification to contact Agilent for your service or support request, please do so. Some instruments cannot have service tickets generated through the CrossLab Connect software.
If you receive an email that your service ticket…
Hi silencer2008,
the 1 µL/s rule means the injection speed used for splitless injections, which is very slow compared to split injection where you want to inject as fast as possible.
Two interesting articles would be:
(+) Pre-Columns - the forgotten…
A follow up question about this, if it ever occurs again. When the instrument errored out it prevented us from even initializing the instrument properly. Is it possible to auto reference even if the unit failed to initialize correctly?
Hi joonseok
Put the samples in a separate sequence to the wash runs. In the sequence parameters for the samples, select the post sequence option to switch off the lamps. Run the sample sequence and queue up the wash sequence from the Run control menu…
Hello, Our lab has a GC-MS (7890B/5977B) system which uses MassHunter version B.07.01 SP1. We recently installed a FID on the system to run FID analysis separate from the MS. We are finding that MassHunter is not as suitable to use for FID analysis in…
I have an agilent GC-FID 6890 i use for TPH and PAH of soil samples yet i want to calibrate it for Oxygenates in gasoline and Aromatics don't know how to go about it
This Information Applies To: HES ion source in 5977B/C GC/MSD, 7010 triple quadrupole GC/MS systems
Issue
The filaments are replaced as part of routine maintenance, or when they break (zero output current).
Items Required
Gloves, nylon,...
This Information Applies To: Agilent 5977A/B/C GC/MSD
Issue
The Electron Multiplier Horn must be replaced when it is contaminated or the EMV value reaches 3000 V with a clean ion source.
Items Required
Electron multiplier horn (5977B/C:...
This Information Applies To: Agilent 5973/ 5975/5977 GC/MSD with rotary vane foreline pump
Issue:
The oil mist filter is recommended to be replaced annually. If the color of the filter has changed or there is smell from the filter, the oil mist...
This Information Applies To: Agilent EI Source
Issue:
During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments...
This Information Applies To: Agilent G7129A/B/C Vialsampler
Issue:
Replacing the needle and needle seat of the Vialsampler is required when:
The needle is visibly damaged.
Leaks or blockages in the vialsampler are observed, especially between...
This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps
Issue:
How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps.
...
This Information Applies To: Agilent UV and RID detectors
Issue
A clogged flow cell causes high backpressure. The clog can be removed by flushing the system with solvents at high pressure. If you suspect a clog or experience high backpressure due...
This Information Applies To: Agilent G1312X and G7112X binary pumps.
Issue:
A blockage before the pressure sensor, e.g. in a pump head, will not give a high pressure or overpressure system error. Instead, the pump may show one of various error messages...
This Information Applies To: Agilent 1100, 1200, 1260 LC Pumps.
Issue:
The following procedure allows the cleaning of residue from the pump's check valves, often caused by polymerization of aged acetonitrile or salt-based solutions that may prevent...
This Information Applies To: Agilent LC Systems
Issue
After analyzing dirty or concentrated samples, some sample compounds may retain in the column and other parts of the HPLC system. If the column re-equilibration after each run is not enough to...