Problem solved. The format in Region settings should be set to "English (United States)", nothing else. As a result, the Decimal symbol under Numbers in Additional settings becomes a point (Anglo-Saxon decimalization) instead of a comma (Western European…
Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
Hello Samehalsharawy ,
Can you please elaborate on the statement we uninstalled the cylinder expecting the flow to come down to zero but it didn't. It stayed at 50?
If no physical tank is connected to the GC, the GC gas connections are capped…
A compilation of our favorite resources to help you get started with your new GC or GC/MS System:
Free Agilent Education courses are available to guide you through site preparation of GC systems including systems with a Mass Spectrometer, Headspace…
This Information Applies To: Agilent MassHunter Qualitative Analysis Software
Issue
Error message prompted “Access to the path C:\Users\AgilentMad\AppData\Local\Temp\FILESPEC.TXT is denied" during search using NIST MS Program in MassHunter Qualitative...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
When sending a run configuration from the AriaMx PC software, no run configuration is transmitted. The expected blue confirmation icon does not appear at the bottom of the touch...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
Can the AriaMx software be used on Windows 11?
Resolution
The AriaMx a nalysis software 2.0 has not been validated for Windows 11, so Agilent...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
How to control more than one AriaMx with one PC.
Resolution
If you connect the PC and instruments using an ethernet cable, you will not be able to establish multiple connections...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
In the Export Data menu, after exporting data in a LIMS format, the output file's name is incomplete.
Background
An example of everything after...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
The AriaMx Data saving behavior when priming experiments from the PC is dependent upon whether the user is logged in or a Guest and whether the user...
This Information Applies To: Agilent OpenLab CDS
Issue
Sequence Creation Template allows users to create a sequence template that identifies the order of injections or actions in the sequence.
This article outlines the step-by-step instruction...
This Information Applies To: Agilent 6100 and 6400 Series LC/MS Systems
Issue
Tuning a mass spectrometer involves optimizing the voltages applied to various optical elements in the ion path to achieve the required mass accuracy, resolution, and...
This Information Applies To: Agilent OpenLab ChemStation
Issue
In OpenLab ChemStation, creating a result set by combining multiple existing result sets can be useful in the following scenarios:
Combining single samples, sequences, or a combination...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
The AriaMx analysis software shows an error when attempting to open a data file.
Background
The AriaMx analysis software can show several errors when something is wrong...