As the filament gets older and the source dirtier areas will go down. Just bump up the EM volts a little by say 24 and see what you get. Keep bumping it up until you get areas you want. It should not change the tune much. You can always set it back to…
Discover the latest Agilent GC/MS and LC/MS technologies . Join thought leaders and your Agilent team of experts for our LC/TQ and IM/Q-TOF user meetings , come to one of our breakfast sessions , or register for a webinar .
We've been having issues with our LC-MS/MS 1260/6490 and 1290/6470 systems giving "LC Offline" errors when the instrument is all ready to go and we start a worklist. It clearly is connected, because it's showing the LC status the entire time. We're using…
From a GCMS perspective, 1.9 ml/min into the ion source is not optimal. The MS prefers down around 1.2 ml/min. Sensitivity is better and spectral tilt is more typical.
Septum head or septumless head? What is your gas supply pressure? When was the last…
Hello,
Question 1: Are you opening a snapshot or using review completed injections to open the in-progress injections? Is it possible you were on a replicate injection when trying to change the type.
Question 2: Andy is correct you must have the…
Helllo mighty Marty,
we have only one Infinity II device, so it can be useful. It works same as your example. But if we are going to compare this system with GC catalog it seems to be very painful to write all the info again and again when you are going…
Hello payninater ,
To utilize the inherited calibration curve you will need to copy the cal levels to the other targets. If you are doing this make certain your unknown targets have the same ISTD assigned. Quant does not verify this and simply applies…
Hello payninater ,
It is not possible to enter responses manually in the Method Editor.
As far as I know there should be no issue with setting your sample type and level in the batch table, creating a new method from acquired data, creating levels…
This Information Applies To: Agilent 4200, 4150 and 2200 TapeStation.
Issue
Is it possible to analyze single stranded DNA (ssDNA) on the TapeStation? Which kit is the most suitable for analysis of ssDNA?
Resolution
Agilent does not offer any...
This Information Applies To: Agilent 4200 and 4150 TapeStation systems
Issue
How to Run the software installation verification on 4200 and 4150 TapeStation systems.
Background
The 4150 TapeStation model G2992A and 4200 TapeStation model G2991B...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, and Femto Pulse
Issue
Perform a capillary alignment when a new array is installed in your Agilent Parallel Capillary Electrophoresis instrument. An alignment may also be performed...
This Information Applies To: Agilent 4200 and 4150 TapeStation
Issue
Are the Agilent DNA and RNA ScreenTape assays denaturing?
Resolution
The DNA ScreenTape assays in the TapeStation portfolio are not denaturing. Only the RNA and High Sensitivity...
This Information Applies To: Agilent LC pumps with manual purge valve.
Issue
A good practice before using a system is to purge the system to remove air. This is done by observing the clear tubing from the solvent bottle as well as the waste line...
This Information Applies To: ICP-MS MassHunter 4.x and 5.x
Issue
Unable to open a batch in offline data analysis, or the batch is empty.
If the ICP-MS MassHunter software closes unexpectedly, there is a power outage, or the ICP-MS and computer...
This Information Applies To: Agilent OpenLab CDS versions 2.4 or higher.
Issue
How to preview result files using various Report Templates in OpenLab CDS versions 2.4 or higher.
Requirements
The user has the result set loaded in OpenLab CDS...
This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data ON.
Issue
An 'Authentication failed' error message appeared after entering user name and password when attempting to download a...
This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data On.
Issue
With Spectroscopy Database Administrator (SDA) and Protect Local Data turned on, batches are not accessible through file...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Scratches on AriaMx optical modules can lead to a reduction in their performance.
Background
Scratches on an AriaMx optical module are shown in Figure 1 .
Figure 1 ....