This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, and Femto Pulse
Issue
Perform a capillary alignment when a new array is installed in your Agilent Parallel Capillary Electrophoresis instrument. An alignment may also be performed during troubleshooting.
Steps to Follow
Tip: An optical alignment can only be performed without dye for a 12- or 48-capillary array.
|
Figure 1. Capillary Alignment Prep window.
Figure 2. Capillary alignment pop-up window. This screenshot is from a 5200 Fragment Analyzer.
Tips: Avoid the Top Red CCD Camera Reference Area and the Capillary Alignment References.
|
Figure 3. Capillary alignment display. This screenshot is from a 5200 Fragment Analyzer.
1. Contrast slider. 2. CCD Camera reference. 3. Box drawn by left-clicking and dragging. 4. Capillary alignment references.
Figure 4. Capillary alignment display. This screenshot is from a Femto Pulse.
1. Contrast slides. 2. Box drawn by left-clicking and dragging. 3. Capillary alignment references.
Icon | Description |
Align | Aligns cursors to peaks. |
Save | Saves changes to the alignment and exits the window. |
Cancel | Cancels any actions and closes the file. |
Auto Space | Auto locates the capillary positions based off the first capillary position. Positions will need manual adjustment. |
Read raw | Opens the Align from File window allowing the user to complete the capillary alignment using a previously run file. |
Table 1. Capillary alignment display menu options.
Learn how to effectively maintain and operate your Agilent Parallel Capillary Electrophoresis systems:
5200, 5300, and 5400 Fragment Analyzer System Manual on Agilent.com
Femto Pulse User Manual on Agilent.com
|