Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-MS Portal
  • Agilent Knowledge Portal
ICP-MS Portal
Articles Batch Will Not Open in ICP-MS MassHunter Data Analysis: Attempt to Recover
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-MS Portal Home
  • ICP-MS Portal Articles
  • ICP-MS Portal Videos
  • More
  • Cancel
  • Created :  28 Sep 2023
  • Modified :  24 Oct 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP-MS Software
  • Product Name :  ICP-MS MassHunter Software
  • Task :  Data Analysis Troubleshooting
Related Articles
  • ICP-MS MassHunter Software Will Not Open or Crashes After Opening

    This Information Applies To: Agilent ICP-MS MassHunter Software 4.x and 5.x Issue ICP-MS Instrument Control or Data Analysis Software will not open, or crashes shortly after opening. Background MassHunter uses configuration files to store software settings and information. Sometimes these files can become corrupt, preventing software operation. Before running the Glitch Resetter (which is a program used to reset all configuration files), it is possible to rename individual files manually...
  • How to Export an ICP-MS MassHunter Batch with SDA and Protect Local Data On

    This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data On. Issue With Spectroscopy Database Administrator (SDA) and Protect Local Data turned on, batches are not accessible through file explorer, and steps must be taken to export/download a batch. Background It may be necessary to export an ICP-MS MassHunter batch from SDA to move a copy of the batch to a new database, share the batch with a colleague or Agilent support for...
  • Creating a New Batch in MassHunter Quantitative Analysis

    This Information Applies To: Agilent MassHunter Quantitative Analysis Software Classic mode, and Agilent MassHunter Quantitative Analysis Software Quant-My-Way mode (Rev. B.09.00 and above) Issue In the MassHunter quantitative analysis, you should begin with creating a new batch file. The following is the procedure to create a new batch. Steps to follow For the MassHunter Quant-My-Way Quantitative software 1. To create a new batch, click File or New Batch (see Figure 1 ). Figure 1...
  • Creating a Method in MassHunter Quantitative Analysis "Quant-My-Way" User Interface

    This Information Applies To: Agilent MassHunter Quantitative Analysis "Quant-My-Way" user interface (version B.09.00 or later) Issue To process the data files in the batch and obtain quantitative results, you must either create a new quantitative method or use a previous quantitative method and then apply it to the batch. Steps to follow 1. Create a new batch and add the applicable data files (Add Samples) to begin the quantitative analysis then initially save the batch. Refer to How...
  • Error Downloading Batch from ICP-MS MassHunter Database Viewer

    This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data ON. Issue An 'Authentication failed' error message appeared after entering user name and password when attempting to download a batch using MassHunter Database Viewer. Background The user name and password entered MUST be that of a windows user who is a member of the windows ICPMHAdmin group. These may not necessarily be the same user name and password used to access the...

Batch Will Not Open in ICP-MS MassHunter Data Analysis: Attempt to Recover

Answer

This Information Applies To: ICP-MS MassHunter 4.x and 5.x


Issue

Unable to open a batch in offline data analysis, or the batch is empty.

If the ICP-MS MassHunter software closes unexpectedly, there is a power outage, or the ICP-MS and computer lose communication during data analysis, the batch may become corrupt. You may face errors when attempting to open the batch in offline data analysis or the batch may open, with no results in the full quant table. 


Steps to follow

Note: There is no guarantee that a corrupt batch can be recovered. The following steps are best effort.

 

Click Here for MassHunter 5.x instructions

Click Here for MassHunter 4.x instructions

 

For MassHunter 5.x:

  1. Open Offline Data Analysis
  2. Create a new batch
    1. On the File Tab (Figure 1; Item 1) click New Batch Folder (Figure 1; Item 2).


      Offline Data Analysis MH 5.x.
      Figure 1. Offline Data Analysis MH 5.x
      1. File Tab, 2. New Batch Folder

    2. Enter a new batch name (Figure 2; Item 1) and click Create Batch (Figure 2; Item 2).


      New Batch window MH 5.x.
      Figure 2. New Batch window MH 5.x
      1. Batch name 2. Create Batch button

  3. Import a DA Method from a 'sister batch' with identical Data Analysis setup - such as analytes and calibration standards/levels.
    1. On the Home Tab (Figure 3; Item 1), click Import DA Method (Figure 3; Item 2).


      Offline Data Analysis MH 5.x.
      Figure 3. Offline Data Analysis MH 5.x
      1. Home Tab 2. Import DA Method

    2. Browse to a batch folder with identical Data Analysis Method setup. It is important to not select the corrupt batch, in case the Data Analysis Method setup is the corrupt component. 
    3. Select the *.batch.bin file (Figure 4; Item 1) and click Open (Figure 4; Item 2).


      Open Method window MH 5.x.
      Figure 4. Open Method window MH 5.x
      1. Batch File 2. Open button

  4. Import All Samples from the corrupt batch.
    1. On the Home tab (Figure 5; Item 1), click Import Samples (Figure 5; Item 2) then Import All Samples From Batch ... (Figure 5; Item 3). 


      Offline Data Analysis MH 5.x.
      Figure 5. Offline Data Analysis MH 5.x
      1. Home Tab 2. Import Samples 3. Import All Samples from Batch ...

    2. Select the corrupt batch (Figure 6; Item 1) and click Open (Figure 6; Item 2).


      Import All Samples from Batch window.
      Figure 6. Import All Samples from Batch window
      1. Batch Folder 2. Open button

  5. Process the batch.
    1. On the Home Tab (Figure 7; Item 1), click Process Batch (Figure 7; Item 2).


      Offline Data Analysis MH 5.x.
      Figure 7. Offline Data Analysis MH 5.x
      1. Home Tab 2. Process Batch

  6. If the batch recovery was successful, you should be able to view all results in this newly created batch.


For MassHunter 4.x:

  1. Open Offline Data Analysis
  2. Create a new batch.
    1. Click the arrow next to Batch (Figure 8; Item 1) and click New Batch Folder ... (Figure 8; Item 2).


      Data Analysis MH 4.x.
      Figure 8. Data Analysis MH 4.x
      1. Arrow next to Batch 2. New Batch Folder ...

    2. Enter a new batch name in the Folder Name field (Figure 9; Item 1) and click Create Batch (Figure 9; Item 2).


      New Batch Folder window.
      Figure 9. New Batch Folder window
      1. Batch name 2. Create button

  3. Import a DA Method from a 'sister batch' with identical Data Analysis setup - such as analytes and calibration standards/levels.
    1. Click the arrow next to Method (Figure 10; Item 1) then click Import DA Method (Figure 10; Item 2).


      Offline Data Analysis MH 4.x.
      Figure 10. Offline Data Analysis MH 4.x
      1. Arrow next to Method 2. Import DA Method

    2. Browse to a batch folder with identical Data Analysis Method setup. It is important to not select the corrupt batch, in case the Data Analysis Method setup is the corrupt component. 
    3. Select the *.batch.bin file (Figure 11; Item 1) and click Open (Figure 11; Item 2).



      Figure 11. Open Method window
      1. Batch File 2. Open button

  4. Import All Samples from the corrupt batch.
    1. Click the arrow next to Batch (Figure 12; Item 1) then click Import All Samples from Batch (Figure 12; Item 2).


      Offline Data Analysis MH 4.x.
      Figure 12. Offline Data Analysis MH 4.x
      1. Arrow next to Batch 2. Import All Samples From Batch ...

    2. Select the corrupt batch (Figure 13; Item 1) and click Open (Figure 13; Item 2).


      Import All Samples from Batch.
      Figure 13. Import All Samples from Batch
      1. Batch Folder 2. Open button

  5. Process the batch.
    1. Click the Process Icon (Figure 14).



      Figure 14. Offline Data Analysis MH 4.x
      1. Process icon

  6. If the batch recovery was successful, you should be able to view all results in this newly created batch.

 

  Learn how to effectively troubleshoot your Agilent ICP-MS System: 
Agilent 7800/7850/7900 ICP-MS with MassHunter
Agilent 8900 Triple Quadrupole ICP-MS with MassHunter e-learning courses available from Agilent education