This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data On.
Issue
With Spectroscopy Database Administrator (SDA) and Protect Local Data turned on, batches are not accessible through file explorer, and steps must be taken to export/download a batch.
Background
It may be necessary to export an ICP-MS MassHunter batch from SDA to move a copy of the batch to a new database, share the batch with a colleague or Agilent support for review, or to use Custom Report Designer. The ICP-MS Database Utility Program is used to download the batch files from the database and restore the files to the batch folder format BatchName(yyyy-MM-dd_HH-mm-ss).b.
Steps to Follow
Figure 1. Starting the ICP-MS Database Utility
1. Start 2. ICP-MS MassHunter Workstation Folder 3. ICP-MS Database Utility
Figure 3. MassHunter Database Viewer
1. Server 2. Database 3. Folder 4. Batch to be extracted 5. Download Batch
Figure 4. Authentication Required Window
Tips: If you receive an error such as "The server could not be contacted" or "The User is not a member of the ICPMHAdmin group", see the article Error Downloading Batch from ICP-MS MassHunter Database Viewer
|
Figure 5. Browse For Folder Window
1. C drive 2. Make New Folder button 3. New folder created and ready to be renamed
Figure 6. Browse For Folder Window
1. Downloaded Batches folder selected 2. OK button
Figure 9. Browse For Folder Window
1. Downloaded Batches folder selected 2. OK button
Figure 10. ICP-MS Database Utility Restore
1. Restore Button 2. Destination of batch
Figure 11. C:\ Download Batches \Download (yyyy-MM-dd_HH-mm-ss)