Hello, H2 is the fuel for the flame in the FID, so if you are out of H2 then the FID will not be able to light. That is very likely the reason why you are not seeing a signal from the FID.
Abdoul-
From your description (instrument configuration fails), it sounds like you will need to remove and re-install MH Acquisition. You might try a repair of the software.
Pinging 192.168.254.12 only confirms that the MS is communicating with…
Thank you very much! I'll make sure to have those items correctly. Our instrument (5975C MSD) has been in idle condition for over an year but under vacuum up until recently I started it again. My real issue is autotune fail. The tune fail symptom is Ent…
I have a 7890 and 5977a which have been sitting for a few years as the result of a failed business venture, I did some successful spme analysis on it a year or so back but just tried some tests from the ALS and it fails. I'm an absolute novice with this…
No, the flows on the two columns downstream of the switch have to be equal. They are feed by the same pressure source, so since MSD is under vacuum and the FID is ambient, the column connected to the MSD has to have more restriction, to keep the flows…
I have decades of experience with consumer products, all on Agilent GC, GCMS, and HPLC.
If I was needing to quantitate 1,2-octanediol in cosmetics samples, I'd extract the sample with dimethylformamide and add BSTFA to make trimethylsilyl derivative…
Your second sample set did not start because there is a conflict that could not be resolved. Instrument failures can be separated into method or sequence.
If you select your (failing) instrument method and press "Setup", does the instrument start equilibrating…
Hello Kang: Generally, and aging lamp would be the likely cause of this problem. The average life of the 8453 lamp is 1000 hours. Lamp performance and lifetime may also be affected by physical shock while the lamp is hot, igniting the lamp while it's…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...