Hi all,
I am currently troubleshooting a leaky 1260 infinity II HPLC-SFC-ELSD and want to carry out a leak test in the utility program, but i could not find a clear, detailed description on how to do this. Could anyone direct me to the correct resource…
Hi!
I still need help yes, please!
- On the machines (GC system and Injector) it just says HP6890 (see photos attached)
- The PC we employ is an old Windows XP running one, that was used to run the GC-FID before we want to control it with the serial…
Hi there, I noticed recently that the ECD back signal on Agilent GC 7890 (using it for trace gas analysis) is oddly spiking (while the FID is a straight, consistent line). It never used to do this. I just recently changed the inlet septum, could this…
Hi, I would suggest to test the same without column installed to the FID. If problem looks similar check for leaks at the FID gas lines and clean the FID. If not clean the inlet and replace the split vent line and filter. Maybe try a new column.
BR…
Hi Community
I have an Issue on one of our GC-FIDs 6890N. I have large Peak Tailing, especially on the back of my chromatogramm:
What i already did:
- New Column
- New Septa/liner
- New Gold Plate
- Cleaned Inlet and Detector
- Re…
Any special procedures to prepare for cross-country shipment a 7890B with SS, PP inlets, heated valve oven, and TCD-FID? Also a 7697A with 100-vial tray?
I have a Cary 3500 and running Cary UV Workstation Software 1.2 and Windows 10
I downloaded the new Cary UV Workstation Software 1.3 via the Agilent SubscribeNet.
first I ran the systemPreparationTool.exe
That finished without issues.
Then I ran…
*The system and method are given at the bottom of this post*
I am working with an Agilent 8860 operating a DB-624 UI (30m x 0.25mm x 1.4um) column with nitrogen carrier gas and split inlet (20:1). The system was shut down for approximately 1 year, and…
Turning off the systems on Friday night that need to be ready to go on Monday morning is not recommended. It can take quite some time for a GC/MS to come back up to operational after being shut down. Even cooling the temperatures all down and then heating…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...