Hi,
this is a very nasty mass axis.
Could you please confirm, that you have S-Lenses installed, as they are configured.
Please have a look, through this troubleshooting guide and follow the instructions. Hopefully it will solve your problems, or…
ICP-MS MassHunter Version 4.1. I cannot Startup Standard Lens Tune!
The message appears "Failed to add Startup/Hardware tasks to Queue; Enter path for batch folder".
As usual (my Batchs)- everything works.
Hello,
Our Headspace sampler 7697A Leak test failed. Leak test failing at the stage 1 off the Restriction and Pressure Decay Test.
Going from there, the sample probe was replaced as well as the pressurization solenoid valve but without success.
…
Hello vtpolywood ,
What type of GC do you have, if I may? If it's an 8890 for example there are some Diagnostic tests built-in that you can run on the Cool On Column inlet. Have you recently changed gas tanks (whether carrier, FID fuel, air or makeup…
In MSD Chemstation, you can specify which signal to use for quantification: TIC, Target Ion, or GC Signal (for GC detectors like FID, TCD, etc.) Go to the Calibrate menu and select Edit
Compounds. On the Compound Table Identification tab, look for the…
Hi,
I plan to attempt some SimDist analysis in accordance with D2887 on my agilent GC-FID. I was hoping that I could achieve this without additional software (i.e. just through manual manipulation of the data acquired by Chemstation), but apparently…
Our micro GC 990 is Natural Gas Analyzer B and B extended version with 3 channels (PoraPLOT U, 6-meter CP Sil 5 CB, CP-Molsieve 5A). We are having trouble with data processing. We have 2 calibration gases from Agilent (Natural Gas mixture and Universal…
Hi everyone.
We have a 8890 Agilent GC with nickel catalyst and FID detector. We would like to use hydrogen as carrier gas, but we don’t know the right flow rates, because the manual is indicating flow rates only for using helium as carrier gas.
For…
the abundance of ion 69. I rechecked the source and it doesn't look dirty. then I did an autotune it stopped on the width mass axis of the second quadrupole.
Finally, after 2 or 3 more attempts, I ended up completing the autotune with an abundance of…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...