This is absolutely the wrong column. Ammonia and H2S need apolar thick films to be separated and no specialized Free-Fatty-Acid-Phase (FFAP). A CP-Volamines column would do (CP7448) . Ammonia elutes just after CO2 and before water, H2S after water at…
Hello,
Does anyone know were the signal data for the live signal displayed in the online signal window is stored and/or how to access it in Chemstation Rev.B.04.03[16]? I would like to write a macro to run in the scheduler to periodically grab a live…
For the GC-6890 FID detector issues you're experiencing, here's a simplified approach to troubleshoot the errors:
Gas Supplies:
Make sure all gas supplies (utility, makeup, hydrogen) are on.
Check that the pressure…
We have a GC-MA7890B with an FID detector and I are observing spikes in the baseline after we made a couple of changes.
We installed guard columns with unions on inlet and detector side, added a 2nd gas cylinder and required tubing, cleaned the FID…
Hi All,
I have an Agilent 7700 ICP-MS running MassHunter 4.6. Interesting issue - the instrument tunes just fine but when I go to run a sample, the data acquisition never stops; it continues well past the time it should have taken including rinse program…
Hi everyone I have a strange issue to report. We are using a 6890GC with FID and TCD that was upgraded many years ago to give use LAN communication rather than GPIB. It works great even though it is over 20 years old. Just prior to Christmas we installed…
Hi,
1. It is not normal, that you have to purge the cell gas lines every time
2. Ou should perform at least following "Tests" during Plasma ignition: Torch Axis - EM - Resolution/Axis. But a performance report every day would be nice (and therefore…
Thank you very much for your quick attention. I take advice to use Labadvisor, however, the host computer is a windows 7 system and the built-in net framework is NE3.5, but LabAdvisor downloaded from Agilent site needs Net Framework version 4.7.2 , which…
Hi All,
We have a 5977 MSD which came up with two errors following a site power outage:
"Source Heater Error: Timeout - temperature fails to increase under full power." "Analyser Heater Error: Timeout - temperature fails to increase under full power…
I am building a report for some GC-FID data from a GC-MS/FID instrument. I have managed to build a report that contains the chromatogram and the table of peaks. I can scale the X-axis to the scale i need but I have been unable to scale the y-axis to show…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...