Hi, I always use P/A factor adjustment under Acq/Setup. That works very well. In all your calibration samples a p/a factor adjustment is done. Takes only a few seconds extra.
We do not maintain listings of spectra in the ST Japan libraries. Please contact ST Japan to obtain a listing of all the spectra in the library. They can reached at Contact - S.T.Japan Europe GmbH (stjapan.de)
When using the Microlab PC SW in conjunction with the 21CFR11 software, the PUBLIC database must be part of the data PATH. It needs to be added to the GROUP (easiest to do it here), the PROJECT or the USER. Otherwise the error described will always be…
Hello:
Please contact your local Agilent office with the serial number of the temperature control box and the instrument. Repair may be an option, otherwise Agilent can quote you on a new one.
Berst regards,
Mike
Hello,
Thank you for reaching out with your VWorks question.
In general, you can either perform this as a series of tasks as you have already done or make use of loops and scripting to condense the protocol. As you noted, using loops will continuously…
On Thursday, June 6, we presented the third installment of our GC Live series. Our speakers discussed how the new single-channel 8850 GC can help labs maximize productivity, efficiency, and uptime.
Watch the recording at your convenience using the link…
Thank you for contacting Cell Analysis Technical Support. This will depend on your instrument type. See this link as well as the screenshot below. Keep in mind that these ranges are pre-normalization.
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...