Hello thank you very much for your response, thanks for the firmware data, the degasser worked perfect, the firmware change no longer, and I can not find the previous version. and are from the 6.80, but how is it possible that a team of thousands of dollars…
Thanks for the reply. I am monitoring at 254 nm. Higher wavelengths do not show that drift. It must be related to the THF (I am using unstabilized) but I do not see how.
Hello, we have an mp aes 4210, we cannot turn it on and we get this error "PLASMA_IGNITION_SHER_GAS_PRESSURE. POP gas pressure out of range". We performed the gas test but it does not pass.
We contacted an Agilent agent and he told us that the problem…
Thanks for the info! What I was thinking about is that some instruments, like GC/MS units sometimes come with recommendations to not put the unit on a UPS due to problems it can cause with the oven. So, I was wondering if Agilent had anything specific…
Hello,
upon start vacuum sequence. You should hear a clik. This is for BK. Monitor IF/BK meter at that point. it has to go lower than 20 Pa. Then you should hear a second click. This time another valve open. It should rise above 20 Pa. Then after…
Hello,
No possible to delete raw data or a run or modify a raw data. You can only modify the method or batch. You can create a new batch an import only the file you wanted to analyze with your batch method.
This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually...
This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems
Issue:
How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA
Caution : All physical...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120...
This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS
Issue:
The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min.
Steps to follow...
This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)
Issue
Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems...
This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)
Issue:
The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.
The PSD enables the...
This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z
Issue:
Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the...
This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP
Issue:
The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use.
The Vapor Generation Accessory...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ...