Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Articles Troubleshooting and Maintenance of AA Systems: Part 1 - Hollow Cathode Lamps
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Atomic Absorption Portal Home
  • Atomic Absorption Portal Articles
  • Atomic Absorption Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Absorption Systems
  • Product Name :  240FS AA 240Z AA 280FS AA 280Z AA 55B AA
  • Task :  Maintenance Troubleshooting
Related Articles
  • Replacing a Hollow Cathode Lamp

    This Information Applies To: Agilent atomic absorption systems (AA 240/280/55B/240Z/280Z) Issue Replacing hollow cathode lamps is a common maintenance procedure and can be used during troubleshooting to isolate the source of an error. Steps to Follow If you are developing a method using an uncoded lamp, install the lamp as described in the following procedure and note the position. The numbers of each lamp position are clearly marked. Enter the lamp position number for the current method...
  • Troubleshooting and Maintenance of AA Systems: Part 3- Graphite Furnace Optimization

    This Information Applies To: Agilent Graphite Furnace Atomic Absorption and Zeeman Instruments (240/280(Z) AA) Issue: How to optimize the Instrument to allow for better results, troubleshooting, after maintenance or lamp replacement. Steps to follow: Warning : The furnace runs at temperatures up to 3000 degree Celsius . Allow furnace to cool completely before touching. Optimize one or more lamps: Lamp optimization is required after cleaning or installing a new lamp...
  • Troubleshooting and Maintenance of AA Systems Part 2B- Flame Optimization

    This Information Applies To: Agilent Flame AA Systems: 55B, 240FS, 280FS Issue: After cleaning and re-installing the flame atomic absorption system components, it is important to optimize the burner position, nebulizer uptake rate, flame conditions, glass impact bead , and mixing paddles . Proper optimization ensures that the atomic absorption system will yield accurate and precise results throughout its lifetime. Steps to follow: Optimize the burner position vertically, horizontally...
  • Flame Optimization for Agilent Atomic Absorption Systems

    This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA) Issue: How to optimize the Agilent Flame AA for better instrument performance and response stability Steps to follow: Use Figure 1 as reference for the procedure on the Flame AA optimization. Figure 1. Agilent Flame AA instrument diagram Tips : Steps 1 through 4 are performed with the flame off in the Optimization Page of SpectrAA by selecting Optimize Lamps ...
  • Removing and Re-Installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD

    This Information Applies To: Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules Issue: Removing and re-installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules. Both the visible and UV lamps are consumable items and will eventually lose signal intensity leading to a drop in detector sensitivity. Also lamps will get to a stage where they can no longer be ignited, causing UV, or Visible lamp ignition errors. Steps to follow 1. Please turn off all lamps...

Troubleshooting and Maintenance of AA Systems: Part 1 - Hollow Cathode Lamps

Answer

 

This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z


Issue:

Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the instrument detection limit. 


Steps to follow:

  1. Check the recommended current and wavelength for the lamp, and check the selections in the software. This information is recorded on the label near the base of the lamp.

  2. On the Analysis page, select Optimize > OK. 

     Tip 
    Tip: Record the gain value the first time you optimize the lamp. Each time you optimize the lamp, the value should be close to the noted value. Track the gain value over time to monitor the performance of the lamp. A significant change in the gain can indicate that the lamp is reaching the end of its operational life.
  3. Slowly rotate the lamp adjustment screws at the base of the lamp, while watching the lamp optimization bar and numerical display. Generally, only small adjustments will need to be made. Record the optimized gain value.

     



  4. If the display goes off-scale, click Rescale and continue optimizing until the maximum value is obtained on the display.

  5. If you have a flame instrument, and you enabled the background correction, you will see a second optimization bar for the deuterium background correction lamp. Simply rotate the deuterium lamp adjustment screws, located at the bottom left of the lamp compartment. Watch the optimization bar and numerical display and adjust the screws to achieve the maximum value. Only minor adjustments will normally be required. Full re-alignment of the deuterium lamp is usually required only when replacing the lamp.

     



 
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help