This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually to ensure optimum performance of the Agilent atomic absorption systems.
Steps to follow:
Schedule of recommended maintenance procedures:
Figure 1. Spray chamber O-rings
1. O-ring on burner assembly 2. O-ring inside the spray chamber 3. O-ring inside the bung
Figure 2: 1. O-rings (x2) that seal the gas inlets to the spray chamber
Cleaning
Tips: Clean contamination immediately using the approved laboratory procedures. If there is contamination inside the instrument, immediately turn off the instrument and call your Agilent representative
|
Caution: Preventing flashbacks: Instrument and consumable maintenance is important to prevent flashbacks or unexpected flame events. See the Safety section on Page 22 of the instrument User's Guide for detailed information on how to prevent flashbacks, and Page 24 for what to do in the event of a flashback.
|
Fuses
Warning: Fire, Hot Surface, and Explosion Hazards: To prevent reduced safety protection or unwanted fusing, always ensure that the code on the fuse cap matches the values shown in the tables next to the fuse holders.
|
The spectrometer contains two operator-accessible fuses located at the back of the instrument. To replace a fuse, disconnect the instrument from the power supply, and replace blown fuses with one of the type and rating indicated next to the fuses.
Tips: Always check the information printed on the back of the instrument for the correct fuse type. The fuses have a code marked on the cap (for example, TF2AH250V). This refers to the fuse characteristic ('T'=time lag, 'F'=fast acting), the current rating ('x' amperes), the breaking capacity ('H'=heavy, 'L'=low), and the voltage rating ('y' volts). This code MUST correspond to the code appearing next to the fuse holders.
|
Learn how to effectively operate your Atomic Absorption System:
Agilent 240/280 Series AA (including Zeeman) User's Guide user manual on Agilent.com
|