Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Articles General Maintenance for the Agilent Flame and Graphite Furnace Atomic Absorption Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Atomic Absorption Portal Home
  • Atomic Absorption Portal Articles
  • Atomic Absorption Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Absorption Systems
  • Product Name :  240FS AA 240Z AA 280FS AA 280Z AA 55B AA
  • Task :  Maintenance
Related Articles
  • Alignment of the Sample Probe on the Agilent PSD 120

    This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System Issue: It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120, so that the sample is injected properly into the graphite tube. The correct alignment of the probe (the end of the capillary protruding from the base of the probe holder) is: The probe locates in the center of the vial when it collects the solution...
  • Replacing the PSD 120 Syringe and Plunger

    This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA) Issue Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems Steps to follow Warning : Sharp Object: The Syringe is glass. If broken during removal or replacement, the pieces of broken glass can cause severe injury Figure 1. Removing the Syringe Block Assembly To Replace the...
  • Troubleshooting and Maintenance of AA Systems: Part 3- Graphite Furnace Optimization

    This Information Applies To: Agilent Graphite Furnace Atomic Absorption and Zeeman Instruments (240/280(Z) AA) Issue: How to optimize the Instrument to allow for better results, troubleshooting, after maintenance or lamp replacement. Steps to follow: Warning : The furnace runs at temperatures up to 3000 degree Celsius . Allow furnace to cool completely before touching. Optimize one or more lamps: Lamp optimization is required after cleaning or installing a new lamp...
  • Atomic Absorption: Resetting Hollow Cathode Lamp Usage Counters

    This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems Issue: How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA Caution : All physical maintenance should be performed with the Flame off and the system cooled Steps to Follow: Open SpectrAA Software and select Administration from the SpectrAA home screen. Select Configuration menu option Usage Counters....
  • AA Calibration Curves Types

    This Information Applies To: Agilent Spectroscopy Systems (AA50,55,240,280) Issue: Typical QA/QC procedures demand calibration through several steps including running interference check samples, calibration verification, running calibration standards, and determination of linear dynamic range. The idealized calibration or standard curve follows Beer's law , which states that the absorbance of an analyte is proportional to its concentration. Deviations from linearity usually occur for various...

General Maintenance for the Agilent Flame and Graphite Furnace Atomic Absorption Systems

Answer

This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)


Issue:

This article includes recommended maintenance procedures to be performed daily, weekly or annually to ensure optimum performance of the Agilent atomic absorption systems.


Steps to follow:

Schedule of recommended maintenance procedures:

  • Daily
    • Check that the exhaust system has positive extraction
    • Check the gas supplies
    • Check all hoses and connections for damage, leaks, and deterioration
    • Empty the drain vessel
    • Clean the Instrument panels/surfaces
    • Clean the burner (Flame systems only)
    • Rinse out the spray chamber and liquid trap (Flame systems only)
    • Check the graphite components-shroud, electrode, tube (Furnace/Zeeman systems only)
    • Clean the electrodes (Furnace/Zeeman systems only)
    • Replace the graphite tube if needed (Furnace/Zeeman systems only)
  • Weekly
    • Check for O-ring deterioration:
      • Three O-rings inside the burner, spray chamber, and the bung (see Figure 1)
      • Two O-rings that seal the gas inlets to the spray chamber (see Figure 2)
    • Clean the lamp and sample compartment windows (wipe away any buildup)
    • Check the air compressor filter
    • Check the water levels in the recirculating supplies/chiller (Furnace/Zeeman systems only)
  • Yearly
    • Arrange for an Agilent service engineer to perform preventative maintenance

Figure 1. Spray chamber O-rings

1. O-ring on burner assembly   2. O-ring inside the spray chamber   3. O-ring inside the bung

 

 

Figure 2:   1. O-rings (x2) that seal the gas inlets to the spray chamber

Cleaning

  • Instrument Surfaces: 
Tip  
Tips: Clean contamination immediately using the approved laboratory procedures. If there is contamination inside the instrument, immediately turn off the instrument and call your Agilent representative
  1.  Switch off the instrument and allow it to cool.
  2.  Remove any hardware present in the sample compartment (see the appropriate instructions in the User’s Guide and the SpectrAA Help if necessary).
  3.  Using a cloth dampened with water, clean the sample compartment, instrument cover, and chimney. On Flame systems, also clean the front panel and flame shield. If necessary, use a mild detergent to clean the cover. Do NOT use an abrasive cleaner. When analyzing biological samples, wipe the instrument with chlorhexidine in 95% ethanol.
  • Windows:
  1.  Clean the sample compartment windows and the lamp compartment window with a gentle stream of filtered air, or a puffer brush, like those used to clean camera lenses.
  2.  If the windows are dirty, wipe them using a lens cleaning paper held between tweezers and dampened with a 50% ethanol/water solution.
  • Filters:
  1.  Check the filter in the air compressor supply line and clean if necessary (instructions are included with the filter).
  • Gas Cylinders:
  1.  Change gas cylinders according to the instructions supplied by the manufacturer.
  2.  When changing gas cylinders:
    1. Inspect all gas supply pipes and hoses. Replace any hose showing signs of damage or deterioration.
    2. Test all connections for leaks using a commercial leak testing solution, a mild dish washing soap solution, or an electronic leak detector.
    3. Test the operation of all regulators.
    4. Ensure that shutoff valves are correctly working.
Caution  
Caution: Preventing flashbacks: Instrument and consumable maintenance is important to prevent flashbacks or unexpected flame events. See the Safety section on Page 22 of the instrument User's Guide for detailed information on how to prevent flashbacks, and Page 24 for what to do in the event of a flashback.
Agilent 240/280 Series AA (including Zeeman) User's Guide


Fuses

Warning  
Warning: Fire, Hot Surface, and Explosion Hazards: To prevent reduced safety protection or unwanted fusing, always ensure that the code on the fuse cap matches the values shown in the tables next to the fuse holders.

The spectrometer contains two operator-accessible fuses located at the back of the instrument. To replace a fuse, disconnect the instrument from the power supply, and replace blown fuses with one of the type and rating indicated next to the fuses.

Tip  
Tips: Always check the information printed on the back of the instrument for the correct fuse type. The fuses have a code marked on the cap (for example, TF2AH250V). This refers to the fuse characteristic ('T'=time lag, 'F'=fast acting), the current rating ('x' amperes), the breaking capacity ('H'=heavy, 'L'=low), and the voltage rating ('y' volts). This code MUST correspond to the code appearing next to the fuse holders.
  • To check a fuse:
  1.  Disconnect the instrument from the main power supply (i.e. unplug the power cord)
  2.  Remove the fuse holder, located to the right of the main input connection. Using a flat bladed screw driver, lever the fuse holder from the instrument.
  3.  Check that the fuses are the correct type and not damaged. If needed, replace the fuses with the correct replacement type.
  4.  Replace the fuse holder in the instrument, and reconnect the instrument to the main power supply.
  • If a fuse repeatedly blows, it may indicate other problems with your Agilent AA instrument. If this happens, contact Agilent Technical Support and arrange for a service.

 

  Tip
Learn how to effectively operate your Atomic Absorption System:
Agilent 240/280 Series AA (including Zeeman) User's Guide user manual on Agilent.com



 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help