Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Articles Replacing the PSD 120 Syringe and Plunger
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Atomic Absorption Portal Home
  • Atomic Absorption Portal Articles
  • Atomic Absorption Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Absorption Systems
  • Product Name :  240Z AA 280Z AA
  • Task :  Maintenance
Related Articles
  • Alignment of the Sample Probe on the Agilent PSD 120

    This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System Issue: It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120, so that the sample is injected properly into the graphite tube. The correct alignment of the probe (the end of the capillary protruding from the base of the probe holder) is: The probe locates in the center of the vial when it collects the solution...
  • General Maintenance for the Agilent Flame and Graphite Furnace Atomic Absorption Systems

    This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace) Issue: This article includes recommended maintenance procedures to be performed daily, weekly or annually to ensure optimum performance of the Agilent atomic absorption systems. Steps to follow: Schedule of recommended maintenance procedures: Daily Check that the exhaust system has positive extraction Check the gas supplies Check...
  • How to Set up the PSD Probe Arm

    This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA) Issue: The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used. The PSD enables the automation of the analysis of samples with the furnace. When in use, the PSD sits in the front sample compartment of the AA, with the carousel containing the required solutions for analysis. When the PSD is set up, it is locked into position to maintain...
  • Troubleshooting and Maintenance of AA Systems: Part 3- Graphite Furnace Optimization

    This Information Applies To: Agilent Graphite Furnace Atomic Absorption and Zeeman Instruments (240/280(Z) AA) Issue: How to optimize the Instrument to allow for better results, troubleshooting, after maintenance or lamp replacement. Steps to follow: Warning : The furnace runs at temperatures up to 3000 degree Celsius . Allow furnace to cool completely before touching. Optimize one or more lamps: Lamp optimization is required after cleaning or installing a new lamp...
  • Troubleshooting a Leak in a Split/Splitless Inlet

    This Information Applies To: Agilent GC Systems with Split/Splitless Inlet. Issue A leak may produce the following symptoms: Inlet pressure unable to reach setpoint (probably a large leak) - will lead to a flow shut-down. Failure of Any Diagnostic Tests from the associated tab on the front panel of the GC and/or the Browser Interface Leak & Restriction Test Pressure Decay Test Septum Purge Test Split Vent Restriction Test The cause of many chromatography issues can be leak...

Replacing the PSD 120 Syringe and Plunger

Answer

This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)


Issue

Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems


Steps to follow

 Warning 
Warning: Sharp Object: The Syringe is glass. If broken during removal or replacement, the pieces of broken glass can cause severe injury

 


Figure 1. Removing the Syringe Block Assembly

 

To Replace the Syringe

  1. Pull the hinged cover downward to access the syringe module.
  2. Carefully pull the syringe block assembly and syringe from the module. See Figure 1.
  3. Unscrew the syringe from the block assembly by removing the two screws at the end of the block holder.
  4. Install the replacement syringe into the block assembly.  Turn the new syringe in the block until it stops moving, but do not over tighten.
  5. Loosen the screws of the syringe block, tighten the syringe by 1/2 of a full turn
  6. Tighten the two screws of the syringe block until they do not freely move.  Do not overtighten.
  7. Replace the syringe block assembly and syringe into the module.  See Figure 2.
  8. Push up the cover to close the module.
  9. Prepare the syringe for use.


Figure 2. Replacing the Syringe Block Assembly

Preparing the Syringe for use

  1. Ensure that the probe is being rinsed properly:
    1. In the SpectrAA Software, in the Furnace Facilities page. select Rinse.
    2. The probe arm should raise and move over the carousel, then return to the rinse vessel.
  2. Once the probe is positioned at the rinse vessel, lift the probe arm in order to clearly view the end of the probe.
  3.  Check for a continuous flow.

To Replace the Syringe Plunger

  1. Repeat Steps 1 and 2 of To Replace the Syringe above.
  2. Remove the plunger from the syringe.  See Figure 3.
  3. Insert the replacement plunger into the syringe.  See Figure 4.
  4. See the tips section for additional articles related to the PSD 120 operation.

 Figure 3. Removing the Plunger

 Figure 4. Installing the Replacement Plunger

 

 Tip 
Tips: See Alignment of the Sample Probe on the Agilent PSD 120 for additional information

 

 Tip 
Tips: See How to Set up the PSD Probe Arm for additional information

 

 Tip 
Learn how to effectively operate your PSD 120:
Agilent Programmable Sample Dispenser PSD 120 User's Guide user manual on Agilent.com

 

 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Privacy Statement
Terms of Use
Contact Us
Site Help