Hello,
You cannot change the presentation of that results preview screen directly in Chemstation. You might be able to change the behavior using macros, but no configuration is available by default.
Marty
Hello,
Based on the ladder profile in the screenshot you shared, it appears that you are running the DNF-905 kit.
In reference to the issue of missing sample signal or blank lanes, since ladder and marker peaks are present as expected, issue is likely…
K_Ruiter The error you are experiencing could be related to the ESW/FW version used on the Ultivo (I would assume you are using ESW 6.74.3). If you are willing to upgrade acquisition software to the latest (MH 12.1), the new ESW/FW for the TQ may resolve…
Hello,
I am not sure as I do not have systems setup that are that old. It would be based on the driver from that version. Do you see the option for serial communication when configuring the HPLC in B.04.03, if not then I would say serial was phased…
HI Arashjafari,
there is a nice overview article here: Chromatographic bioanalytical assays for targeted covalent kinase inhibitors and their metabolites - ScienceDirect
It should work with Poroshell 120 C18, Zorbax RRHD Extend-C18 or ZORBAX Eclipse…
Don't miss our upcoming webinar series in partnership with Dow. We encourage you to join, whether you are an experienced scientist or new to the field, this series is designed to deepen your understanding and provide practical insights.
① Unraveling…
This Information Applies To: Agilent 5000 series ICP-OES systems
Issue
Poor or high relative standard deviation, or a noisy signal is an indication of poor precision during analysis.
Resolution
There are many potential causes for an instrument...
This Information Applies To: Agilent 5000 series ICP-OES systems
Issue
Instrument calibration (wavelength), detector calibration (dark current, or method calibration (worksheet) failure.
Resolution
There are several reasons why a calibration...
This Information Applies To: Agilent 5000 series ICP-OES
Issue
Exhaust temperature and requirements for laboratory site preparation
Background
Certain parameters will impact the actual temperature of the plasma exhaust. Some of these factors...
This Information Applies To: Agilent 700 Series ICP-OES instruments
Issue:
It is important that a torch alignment is performed to ensure that the optics are optimized and are focused on the analytical zone of the plasma. This alignment procedure...
This Information Applies To: Agilent 5800/5900 ICP-OES
Issue:
The instrument requires clean, dry, noncorrosive air for cooling purposes. The air enters the instrument through an air supply vent located beside the chimney on the top of the instrument...
This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES systems
Issue
The ICP-OES spray chamber filters the sample droplets created by the nebulizer. Larger droplets fall to the chamber drain, while the smaller droplets are introduced...
This Information Applies To: Agilent 5000 ICP-OES Series and Agilent ICP Expert Software
Issue
How to connect Agilent 5000 series ICP-OES systems to Agilent ICP Expert software via LAN
Steps to Follow
How to connect Agilent 5000 series ICP...
This Information Applies To: 5000 series ICP-OES (5100/5110/5800/5900), ICP Expert
Issue
How to Reset the AVS 6/7 after maintenance or when a red LED is visible
Steps to follow
Tips : Instructions for changing the AVS switching...
This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES
Issue
The axial pre-optic window needs checking from time to time to ensure it is clean.
If the axial pre-optic window is dirty, it must be cleaned or replaced to ensure adequate...
This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900)
Issue
The ICP-OES cone is the interface between the plasma and the optics via the orifice at the tip of the cone. Argon passes through the orifice of the cone to produce...