Last week we vented and powered off our Agilent 6650 iFunnel QTOF. When we tried to turn it on and pump down from mass hunter software (I think B 09), the rough pump ran at the beginning, but just after 10 min, then it automatically stops. After a couple…
Hello and thanks for your quick answer! I would like to come back to my starting question: if i use macros i can not find the changes from the chemstation in the documentation. For example (one question from above): In Version C.01.10 the Sequence is…
Hi azrael, Thanks for the clarification. If the ground is not functioning at the outlet, the metal of the GC cannot be grounded. This means anything that was touching the metal chassis of the GC could have been subjected to the full line voltage - lots…
I have a G1315C detector that fails "Lowest Intensity Test" in the 190-220 nm range (1867 counts) and the 221-350 nm range (4915 counts). All other intensity values are significantly lower than when conducted a year ago. A new Agilent lamp is installed…
Dear All! A customer would like to do the following (I already advised them about blank subtraction but it is not what they want): How to deduct blank peak area from sample (or standard) peak area What the customer has already done: CC1: to get the blank…
We have Thermo Xcalibur 2.2 to control both Agilent LC1200 stack and Thermo TSQ quantum Ultra, but Xcalibur could not recognize Agilent LC although it can see TSQ MS. it always says " pump/autosampler report a failure when trying to run initialize command…
Hi Marty, Thanks for your reply and suggested CC approach. I will try to clarify what i want the report to do. i have attached an excel file showing an example but it doesn't open properly for me so i have inserted the table below. I want to report the…
Hello, I added tags to this post to increase visibility. It looks like this is a continuation of your previous post: https://community.agilent.com/message/4851-re-automatic-upload-failed-reason-unknown-error#comment-4851 . Is that correct? Typically we…
Hello
I am not an applications expert, but if I had to guess it is because ArC would be present through your calibration - from blank to all standards. ArC present in no gas mode would artificially increases counts in blank and standards, thus potentially…
Hello Renato, Those tests would normally be completed by the Field Service Engineer as part of Operational Qualification for the instrumentation. The engineer has special software tools that enable automated testing and automated Pass/Fail assignment…
This Information Applies To: AriaMx real-time qPCR system
Issue
The instrument fails to reach the required temperature during a run.
Resolution
Check if the temperature of the environment around the instrument is too low resulting in the instrument...
This Information Applies To : Agilent ELSDs G4260A, G4260B, G4261A, G4621B, G7102A, G7102B
Content :
If the evaporator tube of ELSD becomes contaminated with nonvolatile material resulting in poor chromatography, please use the procedure in this...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
The oil should be replaced at least once every six months, or when the oil color is indicates it needs to be changed . The foreline pump should be checked monthly. This article...
This Information Applies To: Agilent ICP-MS MassHunter versions 4.4 - 4.6 and 5.x.
Issue
After performing an autotune, the batch tune report will not generate or print.
Background
ICP-MS MassHunter uses an application called ICP-MS Data Analysis...
This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve
Issue
Installing the Ion Transfer Capillary after cleaning or replacement in Agilent LC/MS instruments with a gate valve.
Required Items...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the procedure on how to quantify uncalibrated compounds in OpenLab CDS 2.x. using reference compounds in the processing method compound table.
Background
Uncalibrated...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the procedure on how to quantify the uncalibrated compounds in OpenLab CDS 2.x when the analytical method requires a fixed response factor.
Background
Uncalibrated...
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
When your Agilent OpenLab ChemStation Edition software fails to launch with errors " Login failed with the following message: Configuration system failed to...
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
Agilent OpenLab ChemStation Edition software fails to launch with error containing the words " Autostart Macro failed ”. This article will provide you some...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software.
Issue Failed to build pressure error comes up shortly after starting a separation...