Hello,
We are having an issue with our 7890B GC back inlet. The inlet temperature and flow were both set to "OFF," but the back inlet kept heating up while were were trying to run the front inlet (in GC-FID mode). This resulted in a thermal shutdown…
H ztarbs,
The air would not be for an inlet supply. This is most likely a detector support gas for FID. Check your method to make sure that the FID is not on in a standby method. FID air consumption is 400 to 450ml/min when on.
Regards
James
We use Agilent NGA Analyzer GC based on 4 column configuration. Hayesep Q pre column, Hayesep Analytical column, molsieve column for TCD and one capillary column for FID. The H2S elution takes place in the Hayesep analytical column. But now a days our…
Hello blairchall ,
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was…
Hello, all you need to do in the method is to uncheck each of the FID flows, and uncheck the electrometer and uncheck the Flame on. With everything unchecked for the FID settings the software will ignore your FID. Under the Signals section you will also…
We've been having issues with our 7890B gas chromatograph. We have it connected to a workstation running Windows 7 SP1 and OpenLab/Chemstation version C.01.07 [27]. We have a Gerstel MPS attached as a headspace sampler.
We have been trying to run…
Hello,
I am getting the following error on OpenLab CDS Chemstation Edition (Rev. C01.10 [287]): "User Change (Unlock) failed, because instrument access is not permitted for user. Logging Out User". We are initially able to login once, but then when…
Hello mdeighan ,
Manual integration is limited to the rules of the selected integrator and the settings of your quant method. The error and behavior you are seeing typically indicate that the peak you are attempting to create is too far outside of the…
This Information Applies To: Agilent ICP Expert software.
Issue
The software version is often asked and is a valuable information: to see if any updates are needed, if certain options are available etc.
Steps to Follow
Launch ICP Expert...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to calculate peak-to-valley ratio in OpenLab CDS.
Background
The peak-to-valley ratio is calculated to indicate the quality of peak separation...
This Information Applies To: Agilent 6500 Series Q-TOF LC/MS systems, MassHunter Workstation 10.1 or below and MassHunter Workstation 11.0 .
Issue
Fully reset the Q-TOF LC/MS system via a power cycle without venting. For users who experience slow...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, and Oligo Pro II systems
Issue
The volume of sample used by Agilent parallel capillary electrophoresis systems during a run.
Resolution
The...
This Information Applies To: Agilent 2100 Expert software, 2100 Expert Security Pack, 2100 Bioanalyzer
Issue
When opening the 2100 Expert software and using the Security Pack, the error message " Start of launcher application failed - The requested...
This Information Applies To: Agilent 6890, 6850, 7890, 8890, 8860, Intuvo 9000 GC systems.
Issue
The exchange of the ignitor in a Flame Photometric Detector (FPD) may be required in the following scenarios:
Preventive or periodic maintenance...
This Information Applies To: Agilent 1290 Infinity Quaternary Pump G4204A, 1290 Infinity II Flexible Pumps G7104A/C, and G7131A/C.
Issue
If the flexible/quaternary pump does not generate flow or build up pressure, it could be due to a stuck inlet...
This Information Applies To: Agilent 5200, 5300 and 5400 Fragment Analyzer systems
Issue
Separation times for the Fragment Analyzer for each array format and kit.
Resolution
Caution : The times in this article are the length of time...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, Oligo Pro II, ProteoAnalyzer systems
Issue
Guidance for installing the Agilent parallel capillary electrophoresis instrument controller and/or...
This Information Applies To: OpenLab CDS (version 2.x) system
Issue:
This article explains how to export OpenLab CDS 2.x data as Archived file.
For copying OpenLab CDS 2.x data to another Projects or another OpenLab CDS systems, you can use data...