Hi Agilent,
Due to helium shortage, we need to change carrier gas for our GC-FID. Most documentation for Agilent suggests either H2 or N2, and the Chemstation configuration shows those options as well.
We are considering using argon, however, I…
Carbon disulfide is another option for wash solvent when running naphtha or crude oil samples. It's response on FID is low compared to hydrocarbons, so there should be little to no response if some residual is in the syringe.
Since the FID and TCD are in series, I would guess that the extra noise on the peak through the FID is due to the TCD valve switching between carrier and reference. Try setting the FID signal rate to 5 Hz. Also, the FID makeup flow can be reduced or turned…
There's a lot going on. You have Column 1 in Constant Flow to a PSD in Constant Pressure with PSD Purge at 0 ml/min. Is there a purged splitter plate installed or a Purged Ultimate Union? The flow into the MS changes throughout the run due to the PSD…
Here is a collection of Agilent LC Resources. Links to user manuals, firmware resources and maintenance videos , application resources , Agilent InterAct! (Virtual LC Simulation) , HPLC Advisor App , LC Supplies and Selector Tools and LC Troubleshooting…
This problem occurred after updating Openlab CDS Software 2.2 Version to 2.4 Version. GC# 7890B With FID , and TCD Detectors With OpenLab CDS Software 2.4 After installing the software version 2.4 and after a gas sample injection I noticed that there…
Dear Colleagues,
I have GC-FID 8890 and i'm using Gas saver feature but I'm observing when I change time of gas saver , the retention time is changed.
for instance, I set gas saver at 20 ml/min after 2 min then inject methanol and the RT is 4.74 but…
The boxes of standards should have shipped with a Certificate of Analysis enclosed. If you go to Agilent.com, you can search on the part number. A number of the hits are CofAs from different batches. part 5080-8858 CofA Lot CF-4631W.pdf (agilent.com)…
Ah, and I think the issue has to do with that I have 2 autosamplers, and 3 columns (MS at back, and 2 different FID columns at front using a 2way splitter). So the MS signal is in the subsequent folder, the FID1 signal is in the correct folder, and FID2…
Hello,
To calculate the RSD and Pass/Fail results, one can try the following:
In the reporting go to the edtor and open "table properties"
In my specific case, I want to calculate RSD for Area, so I choose the column A, pA*s and press "Column…
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...