Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: Agilent MassHunter Qualitative Analysis software version 10.0
Issue
Introduction to MassHunter Qualitative Analysis for Agilent GC/MS – Navigator View.
Steps to follow
Note : Use the demonstration data...
This Information Applies To: All versions of Agilent MassHunter Qualitative Analysis Software , and Agilent MassHunter Quantitative Analysis version B08.00 and up
Issue
Calculation of signal-to-noise in MassHunter is performed using the same parameters...
This Information Applies To: Agilent GC/MS MassHunter Data Acquisition Software
Issue
Preparing and running a sample is essential for data acquisition. The steps below detail how to run a single sample using an Automatic Liquid Sampler (ALS) on...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0
Issue
Introduction to MassHunter Qualitative Analysis for Agilent GC/MS – Compound View.
Steps to follow
Note : Use the demonstration data files...
This Information Applies To: Agilent MassHunter qualitative analysis software
Content:
MassHunter can be configured to label TIC chromatographic peaks with the best match compound name from an MS library search.
Steps to follow:
On the top...
This Information Applies To: Agilent MassHunter GC/MS Data Acquisition (Rev. 10.0 and below)
Issue
Users running MassHunter GC/MS Data Acquisition with Microsoft Windows 10, may encounter warning messages about insufficient memory in the PC when...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software, Revision 7 and up
Issue
This article shows the procedure to change the number of decimal places in the Mass Spectrum Preview and MS Spectrum Results windows of the MassHunter...
This Information Applies To: Agilent OpenLab CDS ver. 2.4 or earlier.
Issue:
If the PC name on a system with OpenLab 2.4 or earlier revisions needs to be changed post installation of the software. This must be done in the right way or the instrument...
This Information Applies To: Agilent OpenLab CDS Software (ver 2.x).
Issue:
This article will give instructions on how to submit a single sample run in the OpenLab CDS Software.
Steps to follow:
Launch the acquisition software through the...
This Information Applies To: Agilent OpenLab CDS products
Issue:
Every computer running Agilent OpenLab CDS Software requires a unique software license for continuous operation. Failure to install a valid license will cause the instrument control...