Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Mass Spectrometry Software Portal
  • Agilent Knowledge Portal
Mass Spectrometry Software Portal
Articles How to Run a Single Sample with ALS in Agilent GC/MS MassHunter Data Acquisition Software
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Mass Spectrometry Software Portal Home
  • Mass Spectrometry Software Portal Articles
  • Mass Spectrometry Software Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Mass Spectrometry Software
  • Component :  GC/MS Data Acquisition Software
  • Product Name :  GC/MSD MassHunter Acquisition Software
  • Task :  Data Acquisition Operation
Related Articles
  • How to Create and Run a Sequence in GC/MS MassHunter Data Acquisition Software

    This Information Applies To: Agilent GC/MS MassHunter Data Acquisition Software Issue Multiple samples can be run without user intervention using an Automated Liquid Sampler (ALS) in combination with a Sequence. Users can perform simple edits to a new or existing sequence template to specify the sequence of vial numbers, sample names, data file names, and storage locations, etc. The sequence can then be run according to the following procedure. Steps to follow The video shown in Figure...
  • How to Align the Agilent 7693A ALS Sample Tray

    This Information Applies To: Agilent 7693A automatic liquid sampler tray Issue: The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully completed. The sample tray has been factory-aligned before shipment. Agilent does not recommend performing this procedure unless the Align Mode light turns on, or if an Agilent support representative request it to be performed. Please do not...
  • Agilent GC Firmware Update Procedure

    This Information Applies To: All Agilent GC systems, Headspace Samplers, and Automatic Liquid Samplers. Issue: Sometimes there maybe a need to upgrade the Agilent GCs, ALSs, or headspace samplers firmware versions to meet the requirement of a software upgrade, Agilent recommendation, or interoperability issues. Steps to follow: Caution : Always use complete downloaded firmware file. Do not interrupt communication, or turn off GC and computer during firmware update process ...
  • Changing the Turret of the 7693A Automatic Liquid Sampler

    This Information Applies To: Agilent 7693A Automatic Liquid Sampler Issue: Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret may need to be replaced when the ALS is exchanged, for replacing the syringe carrier assembly, or to add an autosampler tray to the GC. The standalone turret allows you to analyze up to 16 samples. It holds two solvent positions and one waste bottle...
  • Removing and Installing a Syringe on the Agilent 7693A Autosampler

    This Information Applies To: Agilent 7693A Automatic Liquid Sampler Issue: The automatic liquid sampler may need to be replaced, in case it is worn or dirty. Steps to follow To remove the syringe: Open the injector door. Slide the syringe carriage to the top position if necessary. Completely loosen the plunger thumb screw until it reaches the stop, and lift the plunger carrier off of the syringe plunger. Open the syringe latch by swinging it in a counterclockwise...

How to Run a Single Sample with ALS in Agilent GC/MS MassHunter Data Acquisition Software

Answer

This Information Applies To: Agilent GC/MS MassHunter Data Acquisition Software


Issue

Preparing and running a sample is essential for data acquisition. The steps below detail how to run a single sample using an Automatic Liquid Sampler (ALS) on your GC/MS system.


Steps to follow

The following steps to run a single sample using a previously created method are demonstrated in the video shown in Figure 1.

1. Browse and select the method you wish to use.

2. Select Method > Run Method or click the large green arrow on the Instrument Control panel to display the Start Run dialog box.

3. In the Start Run dialog box, enter a Data File Name.

4. Enter the Sample Name and Misc. Info., which is optional information to be stored with your data file.

5. Enter a Sample Amount, Multiplier, and the Vial Number location where the sample is stored in the ALS tray.

  Tip
Tips: When using the 7693 ALS with a G4515A Barcode Reader (BCR) tray installed, you must enter a value in the Expected Barcode field. If you are not entering a barcode value, enter the non-case sensitive string Off to disable reading of a nonexistent barcode.

6. In the Methods Section to Run box, ensure that Data Acquisition is selected. If you select the Data Analysis(MassHunter DA) which is optional and will start the quantitative data analysis program with the data file acquired after the run.

7. Click OK and Run Method.

Once you submit the single sample run, the vial number, sample name, and data file name are displayed in the sample run box and the run time is displayed above the total run time in the Run Time box.

The status box at the top-left corner shows the system in Pre-Run status, which changes to Run status after the sample is injected.

The Instrument Control panel icons are grayed out and all instrument controls except the Stop Run icon are disabled until the run is over.

 
Figure 1. How to run a single sample using a previously created method

 Tip 
Learn how to effectively operate your Agilent MassHunter Software:
Agilent 5977 GC/MSD with MassHunter Workstation
Agilent 7000 or 7010 Triple Quadrupole GC/MS with MassHunter Workstation e-learning paths available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help