Hi, Here's a screenshot of the system suit, there is no blank subtraction The analyst said he has used the same processing method for a different run with no issues the only difference being that for those runs the MS setting was the same time as the…
When you say that there are no signs of a leak, have you run a pump leak rate test using Agilent Lab Advisor? If not, I would suggest testing this. If you're unlucky enough for your peak to elute just at the point when the outlet ball valve fails to close…
The basic tools I would recommend to have for GC maintenance:
Ceramic column cutter
1/4" wrench (column nuts)
7/16" wrench (gas fittings at back of GC)
1/2" wrench (replace gold seal on S/SL inlet)
Torx 20 driver (removing FID collector assembly…
Agilent GC 7890A, FID, Split Liner. I'm setting up the column to do Hexane injection testing. It was set up this way previously but the program had been deleted and the column removed before I started here. Front Detector/Column which is below baseline…
Hello and Help!
My GC-FID's inlet pressure stays at 0.7 psi and does not reach the setpoint of 10 psi. Makeup gas reaches the 25 psi setting. Liner and septa are fine. I unscrewed the inlet column nut and placed my finger at the bottom of the inlet…
Hello,
My laboratory runs a multilevel calibration with internal standards on a headspace-GC-FID. Our sequence is set up to replace each level of the curve after each calibrator is analyzed and then calculate the concentration of the analyte with the…
I'm going to be calibrating for ethanol on my HS-GC-FID 6890, but I'm having discrepancies between two standards. The standard that I will be calibrating with (Restek's Residual Solvents #1 34105) is 3000ug/mL in N,N-Dimethylacetamide. I'm verifying the…
Hi Dr. Know
Here is some of the information you have requested. Please let me know if you need any further info.
How is your system configured? FID detector using helium as carrier gas, nitrogen as makeup gas.
Which phases are the two columns…
Hi
I am pretty new to GCMS, so may need extra details on how to do things you suggest
I attached a picture of the set up of our gc system, so I can explain better. We recently installed a guard column using 30 cm piece of FS Deactivated 0.250 mm(Agilent…
First of all, thank you for posting publicly, I had not noticed this interface feature of MassHunter Quant, it looks very useful.
Were you able to get it working in 11.1?
If you're allowed to share your "IScriptable" folder, I would be very interested…
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue:
The Flame Ionization Detector (FID) jet could become clogged with ferrule material, column bleed, or heavy molecular analytes. The clogged jet will affect...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different high concentration, split analysis techniques including:
General use liners
Auto inject, general use
Auto...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different splitless trace analysis techniques including:
General use liners
Pressure, temperature, volume (PTV)
Hot...
This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet
Issue
Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet...
This Information Applies To: Agilent GC Systems with Micro-ECD.
Issue
The Wipe Test is an annual test that performed by end user of the micro-ECD (Micro Electron Capture Detector or uECD) to help ensure radioactive safety when using the detector...
This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD)
Issue
Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items...
This Information Applies To: Agilent GC Systems with flame ionization detector (FID)
Issue
Baking out the flame ionization detector at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items required...
This Information Applies To: Micro Electron Capture Detector (uECD)
Issue
Baking the micro electron capture detector (µECD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Warning ...
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue
This article will describe how an operator can configure the makeup gas on both the instrument keypad and the software driver (Method Editor).
Background...
This Information Applies To: Agilent HPLC systems
Issue
The presence of ghost peaks can be a symptom of a system contamination issue, caused by accumulation of sample and/or foreign substances in the flow path. It is critical for high sensitivity...