We have not tested nor validated the use of Oligo Pro II data on ProSize so we cannot comment on why it would work for your colleagues. As for why only a single row is showing, please collect the zip file for your data as outlined in my previous message…
At 310° C ion source temperature, the 502 will always be lower than typical. Also - please reduce the quad temperature down to 150° or maybe as high as 165° C... High quad temperature reduces the lifetime of the quad and only very few applications need…
Hi Andreja,
For calculating area % of two different peaks you must set one baseline left and one baseline right from the peaks. Then you have to enable "Information Pane" when clicking "View" in the menu bar and the results are shown as given in the…
Hi,
Yes it is right, please orientate on the last peak before the impurities start.
The exact baseline settings can be seen when enabling the peak information section below the analysis view (highlighted in green):
Manual integration limits…
We were able to configure the plate, we are good for now.
The 1290 Infinity II we have is coupled to Sciex 6500 MS and is controlled by Analyst Software using the Analyst Device Driver.
Hello,
No, you cannot use classic report styles *.frp from Chemstation in the CDS 2.x software. You can use any IR template built in Chemstation in CDS 2.x by copying the .rdl files over to the new project folders.
Marty
The software (except for MicroLab Pharma) comes with a 90-day trial period before the software needs to be registered. If you get in contact with our sales team, they should be able to get you a trial copy of the software. Find the appropriate contact…
Hello, Is it possible to get the Intelligent rinse working if the rinses occur after the analysis. Let's say I have 2min analysis time and I changed the Probe rinse of samples and stds to 120s and the Probe rinse 1-3 after the initial rinse. Then the…
This Information Applies To: 4150 TapeStation, 4200 TapeStation
Issue
How to deal with a TapeStation Electrode Cartridge change is due warning message.
Background
The Agilent TapeStation controller software will notify users to replace the...
This Information Applies To: Agilent OpenLab CDS
Issue
Sometimes the integration result is not ideal for all injections. This article outlines the step-by-step instruction to manually integrate peaks.
Background
There are 2 ways to integrate...
This Information Applies To: Agilent OpenLab CDS , Agilent OpenLab ChemStation
Issue
Calculation of Area percent or Height percent for uncalibrated compounds in OpenLab software. If all components respond equally in the detector, then Area% and...
This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900)
Issue
How to power off the Agilent ICP-MS Foreline Pump to perform maintenance, or when fully powering off the Agilent ICP-MS System.
Background
The ICP-MS MassHunter...
This Information Applies To: Agilent ICP-MS (7500,7700,8800)
Issue
How to check the E2M18 or E2M18F foreline pump to determine if the pump needs maintenance?
Background
The oil cleanliness and oil level in the E2M18 or E2M18F foreline pump...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to replace the foreline pump oil in the DS 402.
Background
The DS 402 oil should be replaced at least every 3 months or when the oil color indicates it needs to be changed...
This Information Applies To: Agilent 8890/8860 GC Systems
Issue
How to delete data that is stored on the GC instrument.
Steps to follow
Caution : Before deleting data from your GC, Agilent recommends extraction of data as a backup...
This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, Femto Pulse and ZAG DNA analyzer
Issue
When no data is available in the Agilent ProSize data analysis software for one or more capillaries, the capillary(ies) could be misaligned...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to check the MS40+ or MS40-S foreline pump to determine if the pump needs maintenance?
Background
The oil's cleanliness and level, in the MS40+ or MS40-S foreline...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Powering on Agilent ICP-MS Systems after maintenance, long-term shutdown, or for power cycling.
Steps to Follow
Confirm that the ICP-MS power plug...