Generally, you would start method development with what you find in the USP and/or FDA Dissolution Method Databases for volume, media, RPM, etc. requirements. The FDA method database points you to: Dissolution Testing and Acceptance Criteria for Immediate…
That turbo controller and the 300HT turbo pump are obsolete. The RS232 communication kit referenced in the controller manual is also no longer sold by Agilent (previously Varian).
The current model turbo pump of that size is the 305FS turbo. Turbo Pumps…
Hello Kart995,
What vintage is your Bravo, could you please share serial number?
which .dev file are you trying to load did it work on this computer before?
Does the error say "Device type Bravo does not exist"
Are you able to initialize the Bravo…
As the filament gets older and the source dirtier areas will go down. Just bump up the EM volts a little by say 24 and see what you get. Keep bumping it up until you get areas you want. It should not change the tune much. You can always set it back to…
Discover the latest Agilent GC/MS and LC/MS technologies . Join thought leaders and your Agilent team of experts for our LC/TQ and IM/Q-TOF user meetings , come to one of our breakfast sessions , or register for a webinar .
We've been having issues with our LC-MS/MS 1260/6490 and 1290/6470 systems giving "LC Offline" errors when the instrument is all ready to go and we start a worklist. It clearly is connected, because it's showing the LC status the entire time. We're using…
From a GCMS perspective, 1.9 ml/min into the ion source is not optimal. The MS prefers down around 1.2 ml/min. Sensitivity is better and spectral tilt is more typical.
Septum head or septumless head? What is your gas supply pressure? When was the last…
Hello,
Question 1: Are you opening a snapshot or using review completed injections to open the in-progress injections? Is it possible you were on a replicate injection when trying to change the type.
Question 2: Andy is correct you must have the…
Helllo mighty Marty,
we have only one Infinity II device, so it can be useful. It works same as your example. But if we are going to compare this system with GC catalog it seems to be very painful to write all the info again and again when you are going…
This Information Applies To: Agilent ICP-MS MassHunter Workstation 4.6 or below, and ICP-MS MassHunter Workstation 5.1 and above.
Issue
Guide for identifying the current software version your Agilent ICP-MS system is using, to see if any updates...
This Information Applies To: Agilent OpenLab ChemStation, Agilent OpenLab EZChrom, Agilent OpenLab CDS.
Issue
When users of OpenLab CDS forget their passwords, another user with the right privileges will reset their password, however, sometimes...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Error messages related to temperature can cause the plasma to shut off or disable plasma ignition. While a warning message will still allow the plasma to light...
This Information Applies To: Agilent ICP-MS MassHunter Software 4.x and 5.x
Issue
How to view meters in ICP-MS MassHunter Software, or change what meters are selected?
Background
ICP-MS MassHunter allows up to five meters to be displayed at...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to replace the oil mist filter in the MS40+ or MS40-S?
Background
The oil mist filter in the MS40+ or MS40-S should be replaced yearly.
Warning :
...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to check the DS 402 foreline pump to determine if the pump needs maintenance.
Background
The oil cleanliness and oil level in the DS 402 foreline pump should be checked...
This Information Applies To: Agilent ICP-MS systems (7700, 7800, 7850, 7900, 8800, 8900)
Issue
After shutting down and restarting the ICP-MS, sometimes the communication between ICP-MS and PC does not automatically reconnect. It may be necessary...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to replace the Foreline Pump Oil Mist Filter in the DS 402 foreline pump.
Background
The oil mist filter element should be replaced yearly , or when the level of oil in...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil Mist Filter and Odor Element in the E2M18 or E2M18F?
Background
The oil mist filter and odor element should be replaced yearly, or if the...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil in the E2M18 or E2M18F?
Background
The E2M18 oil should be replaced at least every three months or when the oil color indicates it needs...