Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-MS Portal
  • Agilent Knowledge Portal
ICP-MS Portal
Articles DS 402 Foreline Pump Oil Check for ICP-MS Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-MS Portal Home
  • ICP-MS Portal Articles
  • ICP-MS Portal Videos
  • More
  • Cancel
  • Created :  22 Mar 2023
  • Modified :  24 May 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP-MS Systems
  • Component :  ICP-MS Single Quadrupole ICP MS Triple Quadrupole
  • Product Name :  7700 ICP-MS 7800 ICP-MS 8800 Triple Quadrupole ICP-MS
  • Task :  Maintenance
Related Articles
  • DS 402 Foreline Pump Oil Change for ICP-MS Systems

    This Information Applies To: Agilent ICP-MS (7700/7800/8800) Issue How to replace the foreline pump oil in the DS 402. Background The DS 402 oil should be replaced at least every 3 months or when the oil color indicates it needs to be changed. Check the foreline pump oil color and level monthly. Warning : The surface of the foreline pump may be HOT, do not touch the pump until it has cooled. Never add pump oil while the pump is on. Do not touch the oil. The oil...
  • Oil Mist Filter Change for DS 402 Foreline Pump in ICP-MS Systems

    This Information Applies To: Agilent ICP-MS (7700/7800/8800) Issue How to replace the Foreline Pump Oil Mist Filter in the DS 402 foreline pump. Background The oil mist filter element should be replaced yearly , or when the level of oil in the filter casing is above the mark on the oil-level sight-glass. Items Required Oil mist filter for DS 402 (p/n 9499342M002 ) Warning : The surface of the foreline pump may be HOT, do not touch the pump until it has cooled...
  • E2M18 and E2M18F Foreline Pump Oil Check for ICP-MS Systems

    This Information Applies To: Agilent ICP-MS (7500,7700,8800) Issue How to check the E2M18 or E2M18F foreline pump to determine if the pump needs maintenance? Background The oil cleanliness and oil level in the E2M18 or E2M18F foreline pump should be checked monthly. Perform maintenance as needed, based off the monthly check. The E2M18 oil should be replaced at least once every three months, but the type of samples and number of samples analyzed may shorten the life of the oil. For the...
  • MS40-S and MS40+ Foreline Pump Oil Check for ICP-MS Systems

    This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900) Issue How to check the MS40+ or MS40-S foreline pump to determine if the pump needs maintenance? Background The oil's cleanliness and level, in the MS40+ or MS40-S foreline pump, should be checked monthly . Perform maintenance as needed, based off the monthly check. The oil should be replaced at least once every six months , but the type of samples and number of samples analyzed may shorten the life of the oil; adjust the...
  • MS40-S and MS-40+ Foreline Pump Oil Change for ICP-MS Systems

    This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900) Issue How to replace the Foreline Pump Oil in the MS-40+ or MS-40-S? Background The oil should be replaced at least once every - months, or when the oil color is indicates it needs to be changed . The foreline pump should be checked monthly. Items Required AVF Platinum Foreline Pump Oil (p/n X3760-64004 ) Oil Collection Container (at least 1 liter in size) Oil nozzle cap (or funnel) Paper towels 10 mm hex...

DS 402 Foreline Pump Oil Check for ICP-MS Systems

Answer

This Information Applies To: Agilent ICP-MS (7700/7800/8800)


Issue

How to check the DS 402 foreline pump to determine if the pump needs maintenance.


Background

The oil cleanliness and oil level in the DS 402 foreline pump should be checked monthly. Perform maintenance as needed, based off the monthly check. The DS 402 oil should be replaced at least once every three months, but the type of samples and number of samples analyzed may shorten the life of the oil.

  
Warning:
  • Never add pump oil while the foreline pump is on, as it will introduce vaporized pump fluid into the air, which can lead to adverse health effects.
  • Allow the pump and oil to cool, also utilize appropriate Personal Protective Equipment to avoid risk of burns.

 

  
Caution: Only use Agilent Recommended pump oil appropriate for the DS 402 foreline pump. Using a different oil can substantially reduce pump life and invalidates the pump warranty.

Steps to Follow

Checking the Cleanliness of the Oil

  1. You can judge the cleanliness of the oil by the color. Compare the oil color to the chart shown below (Figure 1) and replace the oil if necessary.


    Checking the Oil Color
    Figure 1: Checking the cleanliness of the oil.

Checking the Oil Level

  1. View the oil-level site-glass on the front of the E2M18 foreline pump to visually confirm that the oil level is within the acceptable range when the foreline pump is powered on (Figure 2).


    Checking the Oil Level
    Figure 2: Checking the oil level through the site glass

  2. If the oil level is out of range, visit this article for replacing the oil for instructions on how to add or drain oil to the appropriate level.

 

 
Learn how to effectively maintain your Agilent ICP-MS System:
Agilent 7800/7850/7900 ICP-MS with MassHunter and Agilent 8900 Triple Quadrupole ICP-MS with MassHunter e-learning courses available from Agilent education
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help