Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-MS Portal
  • Agilent Knowledge Portal
ICP-MS Portal
Articles Warning or Error Messages Related to Temperature in ICP-MS Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-MS Portal Home
  • ICP-MS Portal Articles
  • ICP-MS Portal Videos
  • More
  • Cancel
  • Created :  18 Apr 2023
  • Modified :  29 May 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP-MS Systems
  • Component :  ICP-MS Single Quadrupole ICP MS Triple Quadrupole
  • Product Name :  7700 ICP-MS 7800 ICP-MS 7850 ICP-MS 7900 ICP-MS 8800 Triple Quadrupole ICP-MS 8900 Triple Quadrupole ICP-MS
  • Task :  Troubleshooting
Related Articles
  • ICP-OES Interlocks and Internal Monitoring

    This Information Applies To: Agilent 5000 Series ICP-OES Instruments Issue The spectrometer incorporates interlocks and covers that are used to monitor critical operations to ensure Operator and instrument protection. If the instrument is used in any manner not specified by Agilent, this protection may be impaired. 5800 ICP-OES Warning : No interlock or cover must ever be bypassed, damaged or removed. It is good practice to develop safe working habits that do not depend upon...
  • 5000 Series ICP-OES Status Indicators

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue Description of ICP-OES 5000 Series Status Indicators (LEDs) The Agilent 5000 Series ICP-OES (5100/5110/5800/5900) has two different LED indicators to advise the status of the instrument. The state of both these indicators determines if the instrument is in a fault condition or ready to operate condition. The ready to operate condition is shown when both indicators are a steady green color. Any other color...
  • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES

    This Information Applies To: 5100, 5110, 5800, and 5900 ICP-OES instruments Issue The ICP-OES operator can identify the source of plasma ignition failure. The main cause is air entering the plasma. Steps to follow Excessive air entering the plasma. Check the following: The peristaltic pump tubing is clamped and has sufficient clamping pressure. Usually, air is being pulled into the nebulizer because the pump tubing is not properly clamped during ignition. At ignition, the plasma...
  • Routine Maintenance 700 Series ICP-OES Systems

    This Information Applies To: Agilent 700 Series ICP-OES Systems Issue The 700 Series ICP-OES requires regular maintenance, and routine maintenance tasks are able to be performed by the instrument operator. Any maintenance procedures not specifically mentioned in this article should only be performed by Agilent-trained, Agilent-qualified, or Agilent-authorized Field Service Engineers. Warning : Eye Hazard Danger to eyes. The plasma is an intense light source...
  • Troubleshooting an FID not Igniting

    This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000 Issue: This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out. Cause: If the FID is having ignition issues, it is due to the following causes: FID temperature : is too low. Lit Offset : is incorrect. FID gas flows : if the air, hydrogen, or makeup gas flows are not correct (or not meeting setpoints). Gas...

Warning or Error Messages Related to Temperature in ICP-MS Systems

Answer

This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)


Issue

Error messages related to temperature can cause the plasma to shut off or disable plasma ignition. While a warning message will still allow the plasma to light and stay on, it is necessary to correct the issue. 

The troubleshooting procedure in this article is applicable for the following warnings/errors:

  • 1432 Error: Inside housing temperature is too high.
  • 1430 Error: Air intake temperature is too high.
  • 2435 Warning: Inside/Inlet temperature difference is outside the warning limit.
  • 2433 Warning: Inlet temperature is high.

Background

Temperature warnings or errors can be due to insufficient exhaust flow, improper room temperature, insufficient air circulation, or dust accumulation on the back of the ICP-MS.

It is useful to view the Internal Temp and Inlet Temp meters when troubleshooting. See the article: How to Select Meters in ICP-MS MassHunter Software. 
The Internal temperature is also referred to as the Inside temperature. 

Steps to Follow

  1. Check that the sampling cone is tightened
    1. Reinstall the sampling cone to verify the cone is installed properly and retest.
    2. If the issue remains, continue to step 2.
  2. Check the exhaust flow
    1. Power off the ICP-MS system.

        
      Warning: The exhaust system must remain on always while the ICP-MS is powered on in Analysis mode, Standby mode, and Shutdown mode. The next step instructs you to remove the exhaust hose from the ICP-MS, so the ICP-MS system must be powered off before proceeding!
    2. Disconnect the exhaust hose from the chimney of the ICP-MS.

    3. Use an anemometer to measure the exhaust flow from the base of the exhaust hose.

    4. Compare the measured exhaust flow with the specifications in Figure 1.

    5. If the exhaust flow is within specifications, continue to step 3.
      If the exhaust flow is outside specifications, work with your facilities team to increase the exhaust flow, power on the ICP-MS and retest.

      Port Diameter (mm) Exhaust flow
      m3/min m/s
      150 >5 >4.7
      <7 <6.6
      Figure 1. Exhaust Venting Requirements
  3. Check for accumulation of dust on the back of the ICP-MS
    1. Power off the ICP-MS system (if it is not powered off already).
    2. Check the vertical cutouts and circle cutout for the accumulation of dust (Figure 2).
    3. If no dust is present, continue to step 4.
    4. If dust is present, use a vacuum to carefully clean and remove the dust, power on the ICP-MS and retest.


      Rear of ICP-MS
      Figure 2. Rear of ICP-MS

  4. Check the room temperature
    1. Use a thermometer to measure the room temperature at the location of the ICP-MS.

    2. Compare the measured temperature with the specifications in Figure 3.

    3. If the temperature is outside specifications, work with your facilities team to correct the room temperature and retest.

    4. Light the plasma and monitor the room temperature change near the ICP-MS over time. Insufficient air circulation can cause the temperature near the ICP-MS to rise.

    5. Compare the measured temperature change with the specifications in Figure 3.

    6. If the temperature change is outside specifications, work with your facilities team to correct and retest.

       

      Parameter Specification
      Room Temperature 15 to 30 °C (59 to 86 °F)
      < 2 °C/h change and total change should be < 5 °C
      Figure 3. Temperature specifications

  5. If the temperature warnings or errors remain, contact your local Agilent Technical Support.

 

 
Learn how to effectively troubleshoot your Agilent ICP-MS System:
Agilent 7800/7850/7900 ICP-MS with MassHunter e-learning course available from Agilent education

 

Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help