In my case, the problem was due to an account that could not log in to the company-specific drive where all measurement results are saved. If you have the same problem, try another account and it should work.
I will list a few common errors with this. If it isn't one of the common errors, you may need to get tech support involved to fix it because the software may not be working properly.
Is solvent flowing properly? When you tell one line to be used at…
Dear, Sr
When I want to go to execute any program in WinUV software, it gives a message "Error code 9:1 CSM error,CSM error"
It seems a comunnication error, the instrument is connected via GPIB-USB-HS.
Anyone knows how to do?
I assume it is packed tight with insulation? Otherwise I'd just call Agilent for any ideas, maybe a board is bad or something and you need a service call.
This Information Applies To: Agilent ICP-MS MassHunter Workstation 4.6 or below, and ICP-MS MassHunter Workstation 5.1 and above.
Issue
Guide for identifying the current software version your Agilent ICP-MS system is using, to see if any updates...
This Information Applies To: Agilent OpenLab ChemStation, Agilent OpenLab EZChrom, Agilent OpenLab CDS.
Issue
When users of OpenLab CDS forget their passwords, another user with the right privileges will reset their password, however, sometimes...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Error messages related to temperature can cause the plasma to shut off or disable plasma ignition. While a warning message will still allow the plasma to light...
This Information Applies To: Agilent ICP-MS MassHunter Software 4.x and 5.x
Issue
How to view meters in ICP-MS MassHunter Software, or change what meters are selected?
Background
ICP-MS MassHunter allows up to five meters to be displayed at...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to replace the oil mist filter in the MS40+ or MS40-S?
Background
The oil mist filter in the MS40+ or MS40-S should be replaced yearly.
Warning :
...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to check the DS 402 foreline pump to determine if the pump needs maintenance.
Background
The oil cleanliness and oil level in the DS 402 foreline pump should be checked...
This Information Applies To: Agilent ICP-MS systems (7700, 7800, 7850, 7900, 8800, 8900)
Issue
After shutting down and restarting the ICP-MS, sometimes the communication between ICP-MS and PC does not automatically reconnect. It may be necessary...
This Information Applies To: Agilent ICP-MS (7700/7800/8800)
Issue
How to replace the Foreline Pump Oil Mist Filter in the DS 402 foreline pump.
Background
The oil mist filter element should be replaced yearly , or when the level of oil in...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil Mist Filter and Odor Element in the E2M18 or E2M18F?
Background
The oil mist filter and odor element should be replaced yearly, or if the...
This Information Applies To: Agilent ICP-MS (7500/7700/8800)
Issue
How to replace the Forline Pump Oil in the E2M18 or E2M18F?
Background
The E2M18 oil should be replaced at least every three months or when the oil color indicates it needs...