Hi,
Recently Run the Wavelength accuracy and signal to noise at FLD detector for qualification purposes but it not passing. In wavelength accuracy test got Excitation value 350 it ok but in case of emission got 393 nm its failed required minimum 394…
Hi. The pump made loud noises for several days. After that, the pump stopped working normally. Near 600 rpm the pump stops and A-PLUS Software was showing fail status (205) with error code 206 (bit 128) - too hight load. We use Agilent pump - TPS compact…
Csiquet_flex2000 my guess for your 4 signal sources the Save check box is unchecked, so you don't see them as an online signal during acquisition nor as a saved raw data file. The Signal settings under the GC Parameters is only for GC based detectors…
Thanks for advising, do other errors show up in the logbook?
Please watch the movement of the sampler arm during initialisation:
- Is it shaky or jittery? If yes, turn off module and clean transport rods using IPA & lint free cloth.
- Which position…
Hello mr.white, You can do this with a sequence summary report template: Use 'Multi Signal Plot Overlaid' as you've done already; in the properties, the signals need to be overlaid. You may group by signal e.g. FID Front signal/ FID Back signal or, if…
Hello,
We have been having issues with our ICP MS performance and tune reports. Both have been failing with low counts. We have changed the peri-pump tubing, remade all standards, remade the tune solution, replaced the torch, and cleaned the cones…
Hello,
I will assume you mean you are using the backup restore tool for CDS 2.4 not the ECM XT import scheduler. If you have setup an import scheduler task to back the ECM-XT system up that would be self-referential and would not actually back anything…
Hello, I have an issue when I try to install OpenLab CDS EZChrom Version A.04.07 The plan is to install EZChrom on a Citrix to allow the user to retrieve files from ECM for review or reintegration. I start the installation and select the Installation…
Hi dubang69, Im also using export to CDF in OpenLAB 2.3 and OpenLAB 2.2 (with Data Export Addon). Unofrtunatelly second step for these revisions is to open CDF in some open source software as noreuter recommends. Im using lightweight utility UniChrom…
Hi Jennifer We had an GC-MS issue in the past where the rough pump was starting to fail and as a result kept drawing too much current, causing some of the fuses to fail. We also had an issue with a rough pump (on a LC-MS) which was blowing fuses when…
This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. For instructions for removing an ion capillary on LC/MS instruments with a gate valve installed, please click here
Issue...
This Information Applies To: Agilent 6000 series LC/MS instruments
Issue:
Cleaning the ion transfer capillary as part of regular maintenance or as part of troubleshooting when a loss of sensitivity in the instrument occurs.
Steps to follow ...
This Information Applies To: Agilent 6000 Series LC/MS systems.
Issue
Efficient electrospray ionization (ESI) in your LC/MS ion source requires a properly adjusted nebulizer needle in good condition. The nebulizer needle is a consumable item in...
This Information Applies To: Agilent 6000 Series LC/MS systems.
Issue
Checking and adjusting the nebulizer needle is part of regular maintenance or troubleshooting tasks. It is suggested when there is an issue with nebulizer spray shape or spray...
This Information Applies To: Agilent Single Quadrupole LC/MS systems controlled by OpenLab CDS software
Issue
How to check and recognize the current Instrument Status of your 6100 Series and InfinityLab LC/MS systems using OpenLab CDS software....
This Information Applies To: Agilent LC/MS Systems
Issue
Contamination of the tune mix or the Calibrant Delivery System (CDS) of your LC/MS system can cause Autotune failures and drops in sensitivity due to ion suppression.
Items Required...
This Information Applies To: Agilent LC/MS Systems
Issue
When there is suspected contamination of the CDS, low tune response, or possible blockage in the CDS, flushing the Calibrant system may be required.
Warning : This procedure risks...
This Information Applies To: Agilent LC/MS System
Issue
Instructions on how to check the level of calibrant in your LC/MS system, and refill the calibrant bottle if necessary.
Requirements
Checking the level of calibrant should be performed...
This Information Applies To: Agilent 6100 Series LC/MS Instruments
Issue
Instructions on how to fill the calibrant bottle in Agilent 6100 Series LC/MS Instruments when calibrant levels are running low.
Background
The ability to autotune your...
This Information Applies To: Agilent 6000 Series LC/MS systems
Issue
The article details the correct tuning mix to use according to your LC/MS system and the source type installed on your system. In cases where Agilent tuning mixes need to be diluted...