Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC/MS Portal
  • Agilent Knowledge Portal
LC/MS Portal
Articles Nebulizer Needle Replacement in an Agilent Jet Stream or Standard Electrospray Ion Source
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC/MS Portal Home
  • LC/MS Portal Articles
  • LC/MS Portal Videos
  • More
  • Cancel
  • Created :  21 Jan 2023
  • Modified :  9 Jun 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  LC/MS Systems
  • Component :  LC/MS Ion Sources LC/MS Quadrupole TOF LC/MS Single Quadrupole LC/MS Time of Flight LC/MS Triple Quadrupole
  • Product Name :  6220A Time of Flight LC/MS 6224A Time of Flight LC/MS 6230 Time of Flight LC/MS 6410 Triple Quadripole LC/MS 6420 Triple Quadripole LC/MS 6430 Triple Quadripole LC/MS 6460 Triple Quadripole LC/MS 6470 Triple Quadripole LC/MS 6475 Triple Quadripole LC/MS 6490 Triple Quadripole LC/MS 6495 Triple Quadripole LC/MS 6510 Q-TOF LC/MS 6520 Q-TOF LC/MS 6530 Q-TOF LC/MS 6538 Q-TOF LC/MS 6540 LC/Q-TOF 6545 LC/Q-TOF 6545XT AdvanceBio LC/Q-TOF 6546 LC/Q-TOF 6550 iFunnel LC/Q-TOF 6560 Ion Mobility LC/Q-TOF InfinityLab LC/MSD InfinityLab LC/MSD XT InfinityLab LC/MSD iQ Ultivo Triple Quadrupole LC/MS
  • Task :  Maintenance
Related Articles
  • What to do if my LC-QQQ Autotune Fails?

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems Issue Autotune fails to complete. Background Agilent 6400 LC/MS systems use an Autotune algorithm to optimize the sensitivity and mass resolution of the instrument. Sometimes when running the autotune procedure the autotune does not complete and the user is given an Autotune failed message in MassHunter. One of the main reasons for Autotune failure is a lack of sensitivity or abundance of the tuning ions...
  • How to Check and Adjust the Nebulizer Needle on Agilent LC/MS Systems

    This Information Applies To: Agilent 6000 Series LC/MS systems. Issue Checking and adjusting the nebulizer needle is part of regular maintenance or troubleshooting tasks. It is suggested when there is an issue with nebulizer spray shape or spray consistency, leading to a loss of sensitivity in the instrument. Signs the nebulizer needle requires adjustment or replacement include: Increased LC backpressure associated to the MSD. Off-axis spraying. Dripping from the nebulizer. Reference...
  • Checking for a Blocked ICP-OES Nebulizer

    This Information Applies To: Agilent 5000 Series ICP-OES Issue Blocked nebulizers are the source of many instrument performance issues including loss of sensitivity and poor precision, especially when running organics, high solids, or high salt concentration solutions. The ICP-OES backpressure Nebulizer test verifies nebulizer performance. Run this test if you notice no mist in the spray chamber, have low sensitivity during analysis, or high % RSDs on your replicates. Blockages and partial...
  • Poor Precision in ICP-OES

    This Information Applies To: Agilent 5000 series ICP-OES systems Issue Poor or high relative standard deviation, or a noisy signal is an indication of poor precision during analysis. Resolution There are many potential causes for an instrument exhibiting poor precision, including but are not limited to: V-groove nebulizer Concentric glass nebulizer Torch Contamination in spray chamber Pump/pump tubing Air leak Instrument calibration Aqueous analysis after non-aqueous...
  • Calibration Failure Troubleshooting on ICP-OES Systems

    This Information Applies To: Agilent 5000 series ICP-OES systems Issue Instrument calibration (wavelength), detector calibration (dark current, or method calibration (worksheet) failure. Resolution There are several reasons why a calibration may fail. The following is a list of common areas to check if problems are encountered with calibrations. The steps to follow will depend upon whether the calibration tests start, whether all the lines in your calibration have failed, or whether...

Nebulizer Needle Replacement in an Agilent Jet Stream or Standard Electrospray Ion Source

Answer

This Information Applies To: Agilent 6000 Series LC/MS systems.


Issue

Efficient electrospray ionization (ESI) in your LC/MS ion source requires a properly adjusted nebulizer needle in good condition. The nebulizer needle is a consumable item in your ion source that may need to be replaced when the following signs are observed:

  • Increased LC backpressure associated to the MSD.
  • Off-axis spraying
  • Dripping from the Nebulizer
  • Poor or inconsistent spray shape

This article describes how to replace the nebulizer needle in the ion source of your Agilent LC/MS system.


Required Items

  • Clean powder free nitrile gloves
  • Nebulizer Adjustment Fixture (Part of the Nebulizer Adjustment Kit, p/n G1960-67470)
  • T6 Torx driver (Agilent p/n 8710-2548)
  • 2 x 1/4-inch x 5/16-inch open-end wrench
  • Nebulizer needle replacement kit - select one of the following depending on the ion source type installed on your instrument:
    • G1946-68703 - Replacement needle kit or standard ESI sources, not compatible with Agilent Jet Stream (AJS) sources
    • G1958-60137 - Replacement needle kit for AJS Nebulizer, not compatible with standard Agilent ESI sources

Steps to follow

  1. Put on clean powder free nitrile gloves

  2. Determine the nebulizer holder type by comparing your nebulizer to Figure 1.


    Figure 1
    Figure 1. Nebulizer holder type
    Left - Nebulizer with locknut; Center - Nebulizer with Torx screw; Right - Nebulizer with hex screw and 3 mm wrench

      
    Caution:  When inserting the nebulizer into the nebulizer adjustment fixture, ensure that the nebulizer tip does not contact the fixture, or the nebulizer needle tip may be damaged.

     

  3. Mount the nebulizer assembly into the nebulizer adjustment fixture as shown in Figure 2. 
    Slide the heads of the two spring loaded socket head cap screws through the mounting holes of the nebulizer.
    Twist the nebulizer clockwise to hold the nebulizer assembly firmly in the fixture.
    Gently tighten these screws further to hold the nebulizer firmly in the holder.



    Figure 2. Inserting and tightening the nebulizer assembly into the nebulizer adjustment fixture
  4. Loosen the locknut next to the zero-dead-volume (ZDV) union and then remove the union from the nebulizer needle holder, as show in Figure 3.



    Figure 3. Loosening the lock nut and removing the union from the nebulizer.
  5. Remove the union locknut from the nebulizer.

  6. Loosen the fastener (locknut, 1.5 mm hex screw, or T6 Torx screw that secures the needle holder in place. (See Figure 1)

  7. Loosen the needle holder and pull it out of the nebulizer body.
    Dispose of the used needle safely according to your laboratory disposal protocols. See Figure 4.



    Figure 4. Removing needle and needle holder from the nebulizer body.
  8. For a more thorough cleaning, remove the nebulizer body from the adjustment fixture block, then sonicate the nebulizer body in LC/MS-grade isopropanol (IPA) or methanol for a least 15 minutes.
    You can also pour the cleaning solvent through inside the nebulizer shaft.
    Dry the nebulizer body completely and then reinstall on the adjustment fixture block.

     

      
    Caution: Do not bend the needle in the next steps as you slide the needle into the ferrule. The needle is fragile, and a bent needle results in poor alignment with the nebulizer body.

     

  9. Carefully remove the new needle from the shipping tube and identify the sharp and blunt end of the needle. See Figure 5.

  10. Slide the blunt end of the needle through the needle holder and into a new black ferrule until 1 cm of the blunt end of the needle sticks out of the ferrule. Then push the needle back into the needle holder until the blunt end of the needle is flush with the end of the ferrule. See Figure 5 for more detail.



    Figure 5. Inserting the blunt end of the new needle into the needle holder and ferrule.

  11. Reinstall the locknut and the union, then hand tighten the union onto the end of the needle holder.

  12. Hold the needle holder steady on a clean surface with either a 3 mm open-end wrench or 1/4-inch x 5/16-inch open-end wrench (depends on your nebulizer holder style),
    Tighten the union 1/2 to 3/4 of a turn, or until an audible squeak is heard. The ferrule is now compressed onto the needle, see Figure 6.



    Figure 6. Installing the lock nut and union onto the needle holder, then compressing the ferrule onto the needle.
  13. With clean gloves, pull carefully on the needle to ensure that the needle is held firmly in place by the compressed ferrule.
    Take care of the sharp end of the needle when performing this check.

  14. Insert the needle and needle holder into the nebulizer body and turn the needle holder to engage the thread in the nebulizer body, see Figure 7.



    Figure 7. Installing the needle holder into the nebulizer body.
  15. Adjust the new nebulizer needle to the correct position by following these instructions: How to Check and Adjust the Nebulizer Needle on Agilent LC/MS Systems

 

  Learn how to effectively operate and maintain your Agilent LC/MS system
Essential Housekeeping Tasks and Best Practices for Agilent 6400 Series LC/TQ e-learning course available from Agilent education
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Privacy Statement
Terms of Use
Contact Us
Site Help