Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC/MS Portal
  • Agilent Knowledge Portal
LC/MS Portal
Articles Checking the Instrument Status of your Agilent Single Quadrupole LC/MS System Using OpenLab CDS
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC/MS Portal Home
  • LC/MS Portal Articles
  • LC/MS Portal Videos
  • More
  • Cancel
  • Created :  31 Jan 2023
  • Modified :  9 Jun 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  LC/MS Systems
  • Component :  GC/MS Single Quadrupole
  • Product Name :  6130 Single Quadrupole LC/MS 6150 Single Quadrupole LC/MS InfinityLab LC/MSD InfinityLab LC/MSD XT InfinityLab LC/MSD iQ
  • Task :  Operation Troubleshooting
Related Articles
  • Venting and Shutting Down your Single or Triple Quadrupole LC/MS System

    Multiple Results
    This Information Applies To: Agilent 6100 and 6400 Series LC/MS Systems controlled by OpenLab CDS ChemStation or MassHunter Workstation for LC/QQQ Issue How to safely vent and turn off your Agilent Single or Triple Quadrupole LC/MS system. ...
    • Found In:
    • Articles
    • ,
    • Articles
  • Requesting Tune Control on your GC/MSD System Using OpenLab CDS

    Multiple Results
    This Information Applies To: Agilent Single Quadrupole GC/MS systems controlled by OpenLab CDS. Issue The OpenLab CDS user needs to request tune control before any tuning or vacuum control of the GC/MSD can be performed. Steps to Follow ...
    • Found In:
    • Articles
    • ,
    • Articles
  • Why is my Single Quadrupole LC/MS Not Ready?

    This Information Applies To: Agilent 6100 series Single Quadrupole (SQ) LC/MS Instruments using OpenLab CDS software Issue Waiting to run a sample or tune the instrument but the MS is shown as not ready. Background The MS requires accurate temperature, gas flow, and electronic control in order to maximize performance. If one of the MS setpoints is not yet reached, or has not stabilized, then the MS is shown as "Not Ready" and cannot analyze samples or tune. This status is to prevent the...
  • Put your Single Quadrupole LC/MS Instrument into Standby Using OpenLab CDS

    This Information Applies To: Agilent 6100 Series LC/MS instruments controlled by Agilent OpenLab CDS software Issue Instructions for putting your LC/MS instrument into Standby mode after you have finished analysis. Requirements The LC/MS should have finished all analysis in the sequence. Steps to Follow From the top ribbon, in the Home tab, click Method > Instrument Status. Figure 1 . Instrument Status location in the top ribbon If necessary, in the Instrument section of...
  • Turn off your Single Quadrupole LC/MS System Using OpenLab CDS

    Multiple Results
    This Information Applies To: Agilent 6100 Series single quadrupole LC/MS systems controlled by OpenLab CDS Software Issue Instructions on how to safely turn off the Agilent 6100 Series or InfinityLab Series LC/MS System Steps to Follow ...
    • Found In:
    • Articles
    • ,
    • Articles

Checking the Instrument Status of your Agilent Single Quadrupole LC/MS System Using OpenLab CDS

Answer

This Information Applies To: Agilent Single Quadrupole LC/MS systems controlled by OpenLab CDS software


Issue

How to check and recognize the current Instrument Status of your 6100 Series and InfinityLab LC/MS systems using OpenLab CDS software. 


 

Steps to Follow

  1. If necessary, in the top ribbon, click the Home tab, the click Instrument Status.

    Figure 1
    Figure 1. Instrument Status option location

  2. In the Instrument Status window, click the Dashboard tab to see the SQ Instrument Status pane

    Figure 2
    Figure 2. SQ Instrument Status Pane

The icons in the SQ Instrument Status pane have the following meaning and function:

 

The (+) icon indicates that all the rows in the Acquisition Parameters table in the current method have a Positive polarity.
If the icon alternates between positive and negative, then the current method has at least one row with positive polarity and one row with negative polarity.

 

The (-) icon indicates that all the rows in the Acquisition Parameters table in the current method have a Negative polarity.
If the icon alternates between positive and negative, then the current method has at least one row with positive polarity and one row with negative polarity.

 The type of ion source is shown next to this icon, in this example is an ESI source running in positive mode. If the source is a multimode source (MMI), then the Ion mode (ESI, APCI, or ESI+APCI) is also shown.

 

This icon indicates that the LC stream is directed to the mass spectrometer.

 

This icon indicates that the LC stream is being directed to waste.

 

This icon indicates whether the Calibrant in on or off. If the icon is coloured, then the calibrant is turned on.

 

If the icon is unavailable, then the Calibrant is off

You can turn the Calibrant on or off by right-clicking in the SQ instrument status pane and then selecting the Calibrant command in the shortcut menu.

 

Possible Actions from the SQ Instrument Status Pane

Right-clicking the SQ Instrument Status Pane brings up a shortcut menu with MSD control options. (See Figure 3)



Figure 3. Actions from SQ iInstrument Status pane 

See the following able for a description of the options.

On

This command sets the Single Quadrupole in a ready state. It will be is ready to start a run. The Gas Flow, Nebulizer, and the temperatures are set to the setpoints in the current method. Voltages are applied to the Instrument.

Standby 

Sets the MS to Standby state, where it is not ready to run while saving Nitrogen.


  • Quad Temperature is set to 100 °C.
  • Drying gas heater temperature is set to 300 °C.
  • Nebulizer pressure is set to 20 psig.
  • Drying gas flow is set to three Standard Litres/Minute (SLPM) nitrogen.
  • "Stream" selector valve set to LC > Waste
  • If an APCI source is installed, vaporizer gas temperature is set to 325 °C.
  • If an APPI source is installed, you can turn the APPI lamp on or off.
Calibrant - On or Off

Turns on, or off the calibrant flow to the MS.

LC - To MS or Waste

Changes the direction of the LC stream to the mass spectrometer, or to waste.

When the calibrant is turned on, the LC stream is diverted to waste

Vent 

Vents the MS to atmospheric pressure. A warning message will be given when this option is selected because it can take several hours to re-establish the proper vacuum levels and to re-equilibrate temperatures in the case of an unwanted Vent of the MS.

For more information on safely venting and shutting down your LC/MS systems see the following articles: 
Venting and Shutting Down your Single or Triple Quadrupole LC/MS System
Turn off your Single Quadrupole LC/MS System Using OpenLab CDS

Pump Down

Pumps down the MS system to achieve high vacuum. When the pump down process finishes, the MS is in the Standby state. You can monitor the vacuum status in the SQ tab in the Instrument Status window and check Rough Vac and High Vac values.

For more information on pumping down your LC/MS systems see the following article: 
How to Pump Down your Single Quadrupole LC/MS System Using OpenLab CDS

Schedule Tune 

Opens the Schedule Tune dialog box so that you can schedule an Autotune or Check tune on your instrument.

Note: This feature is only available on Agilent LC/MSD iQ instruments.

Maintenance

Opens the Maintenance dialog box so that you can enable or disable some of the Early Maintenance Feedback (EMF) counters.

Note: This feature is only available on Agilent LC/MSD iQ instruments.

Review Tune Report

Opens a dialog box to allow you to choose a tune file from a list of Autotune or Check tune files. You can then review that Autotune or Check tune report.

 

 

  Learn how to effectively operate and maintain your Agilent LC/MS System:
Agilent LC/MS page on OpenLab Help and Learning
Agilent OpenLab CDS for InfinityLab LC/MSD Series and 6100 Series LC/MS Quick Start Guide user manual on Agilent.com

Agilent Single Quadrupole LC/MSD with OpenLab CDS e-learning path available from Agilent education