Hello, H2 is the fuel for the flame in the FID, so if you are out of H2 then the FID will not be able to light. That is very likely the reason why you are not seeing a signal from the FID.
I have a 6490 QQQ. Each time I run either autotune or check tune, I get masses that are out of tolerance especially in MS2. The system seems to pass if the Dynamic Ramp Table enable button is checked. This is associated with the collision cell and detector…
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was?
From what I understand…
Based on your latest results, I would say that nothing is fundamentally wrong with the optical unit in terms of noise, so I believe this is more an applicational problem - your passed quick noise test points to that direction. The error 1050 with the…
No, the flows on the two columns downstream of the switch have to be equal. They are feed by the same pressure source, so since MSD is under vacuum and the FID is ambient, the column connected to the MSD has to have more restriction, to keep the flows…
I have decades of experience with consumer products, all on Agilent GC, GCMS, and HPLC.
If I was needing to quantitate 1,2-octanediol in cosmetics samples, I'd extract the sample with dimethylformamide and add BSTFA to make trimethylsilyl derivative…
Disconnect the sample probe from the valve and start the leak and restriction test. If it passes this step, it means that there is some restriction in the sample probe, enough to cause that part of the test to fail. Try a brand new probe. If it fails…
Seems then to indicate something in the flow path when including the mixer and damper. First, make sure that you do an elasticity calibration because you have changed important parts in the flow path. Then re-run the pump leak rate test with the mixer…
Hello everyone,
We have a problem with the acquisition methods on GC, but not on HPLC.
When we want to change an acquisition method that is in "approved" status to "generic" status in order to modify it, it is impossible to do so because the box is…
This Information Applies To: Agilent 6000 Series LC/MS systems.
Issue
Checking and adjusting the nebulizer needle is part of regular maintenance or troubleshooting tasks. It is suggested when there is an issue with nebulizer spray shape or spray...
This Information Applies To: Agilent Single Quadrupole LC/MS systems controlled by OpenLab CDS software
Issue
How to check and recognize the current Instrument Status of your 6100 Series and InfinityLab LC/MS systems using OpenLab CDS software....
This Information Applies To: Agilent LC/MS Systems
Issue
Contamination of the tune mix or the Calibrant Delivery System (CDS) of your LC/MS system can cause Autotune failures and drops in sensitivity due to ion suppression.
Items Required...
This Information Applies To: Agilent LC/MS Systems
Issue
When there is suspected contamination of the CDS, low tune response, or possible blockage in the CDS, flushing the Calibrant system may be required.
Warning : This procedure risks...
This Information Applies To: Agilent LC/MS System
Issue
Instructions on how to check the level of calibrant in your LC/MS system, and refill the calibrant bottle if necessary.
Requirements
Checking the level of calibrant should be performed...
This Information Applies To: Agilent 6100 Series LC/MS Instruments
Issue
Instructions on how to fill the calibrant bottle in Agilent 6100 Series LC/MS Instruments when calibrant levels are running low.
Background
The ability to autotune your...
This Information Applies To: Agilent 6000 Series LC/MS systems
Issue
The article details the correct tuning mix to use according to your LC/MS system and the source type installed on your system. In cases where Agilent tuning mixes need to be diluted...
This Information Applies To: Agilent 7693A, 7650A, and 7683 Series automated liquid samplers
Issue
GC software or GC LED display shows ALS Failure code 245: Autosampler front tower plunger error or ALS Failure Code 945: Auto injector back tower...
This Information Applies To: TapeStation ScreenTapes and reagents
Issue
Using the search tool on Agilent.com to locate the declarations of conformity (DOC) for Agilent ScreenTapes and Test-Tapes, certificates of conformity (COC) for TapeStation...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The Automatic Liquid Sampler (ALS) system calibration process aligns the sample tray with the injector turret position to enable vial transfers without incident.
Calibration...