Hi everyone,
Our lab own a microGC 990 with a PORAPLOT U column. The software we use is Openlab.
I usually run my samples with "sequence" mode. Is there a way to put a "delay time" between samples? I just want a delay between finishing sample 1 and…
In the Offline Worklist Editor for MH12, when using the "Edit Samples" window to input the sample positions, the "Current Configuration" box in the top left isn't shaded out, so that I can't choose it. When using Acquisition to do this same function,…
5 psi inlet pressure running hydrogen with your configuration leads to 2.3 ml/min to the MS and 1.02 ml/min to the FID.
You are welcome. Are you using a two-hole ferrule in the injection port or a splitter of some type? The two-holed ferrules are..…
Richard987 A normal EPC module could probably supply the nitrogen to the collision cell but it's not tested, not supported, no one knows as it's never been done. If you can run it now and MassHunter Acquisition isn't complaining it might work.
How long…
Which liner is being used? Check out: GC Liner Selector | Agilent You should think about trying one with a frit instead of glass wool, like this one: 5190-5105 | Agilent Or a single tapered, ultra inert liner 5190-2295 | Agilent .
Splt/Splitless Inlet…
Hi JPhillips - in our latest software (QTOF Acq 12.1) - it is much easier to copy and paste the entire "Sample Name" column into the data file column. All rows in the data file column will automatically receive the .d suffix in one action (the paste command…
Hello GUX ,
For these types of errors the first thing to try is resetting the user preferences for Profinder. To do this, first close all copies of Profinder and then navigate to the folder
C:\Users\XXX\AppData\Roaming\Agilent_Technologies,_Inc\ …
The error means that the maximum stroke length of your pump drive is being shorter than it should be.
This can be either due to mechanical issues like a broken piston or an offset in the pump drive gear leading to a physical block before the normal…
Hello aafzal3 ,
I apologize for the delayed response.
You can use either find by method to attempt to find your targets, but to troubleshoot the issue you need to find your target manually by extracting the EIC for the expected molecular ion. Once…
This Information Applies To: Agilent 6000 Series LC/MS systems.
Issue
Checking and adjusting the nebulizer needle is part of regular maintenance or troubleshooting tasks. It is suggested when there is an issue with nebulizer spray shape or spray...
This Information Applies To: Agilent Single Quadrupole LC/MS systems controlled by OpenLab CDS software
Issue
How to check and recognize the current Instrument Status of your 6100 Series and InfinityLab LC/MS systems using OpenLab CDS software....
This Information Applies To: Agilent LC/MS Systems
Issue
Contamination of the tune mix or the Calibrant Delivery System (CDS) of your LC/MS system can cause Autotune failures and drops in sensitivity due to ion suppression.
Items Required...
This Information Applies To: Agilent LC/MS Systems
Issue
When there is suspected contamination of the CDS, low tune response, or possible blockage in the CDS, flushing the Calibrant system may be required.
Warning : This procedure risks...
This Information Applies To: Agilent LC/MS System
Issue
Instructions on how to check the level of calibrant in your LC/MS system, and refill the calibrant bottle if necessary.
Requirements
Checking the level of calibrant should be performed...
This Information Applies To: Agilent 6100 Series LC/MS Instruments
Issue
Instructions on how to fill the calibrant bottle in Agilent 6100 Series LC/MS Instruments when calibrant levels are running low.
Background
The ability to autotune your...
This Information Applies To: Agilent 6000 Series LC/MS systems
Issue
The article details the correct tuning mix to use according to your LC/MS system and the source type installed on your system. In cases where Agilent tuning mixes need to be diluted...
This Information Applies To: Agilent 7693A, 7650A, and 7683 Series automated liquid samplers
Issue
GC software or GC LED display shows ALS Failure code 245: Autosampler front tower plunger error or ALS Failure Code 945: Auto injector back tower...
This Information Applies To: TapeStation ScreenTapes and reagents
Issue
Using the search tool on Agilent.com to locate the declarations of conformity (DOC) for Agilent ScreenTapes and Test-Tapes, certificates of conformity (COC) for TapeStation...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The Automatic Liquid Sampler (ALS) system calibration process aligns the sample tray with the injector turret position to enable vial transfers without incident.
Calibration...