Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: Agilent OpenLab CDS
Issue
Identifying the baseline codes and baseline code descriptions for OpenLab CDS
Background
OpenLab CDS Data Analysis includes the same integration algorithms as available in both OpenLAB...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to filter a signal in an Intelligent Report.
Requirements
To be able to carry out the procedure as described, the privilege of Data Processing...
This Information Applies To: Agilent OpenLab CDS
Issue
This article outlines different ways to quantify the amount of compound in an unknown sample.
Background
Quantitation uses peak area or height to determine the amount of a compound in...
This Information Applies To: Agilent G1329A/B and G1313A Standard Autosampler
Issue
Replacing the needle seat of the Agilent 1100/1200/1260 Standard Autosampler may be required during regular maintenance, replacing a bent needle, or flow path blockage...
This Information Applies To: Agilent G1329A/B and G1313A Standard Autosampler
Issue
Replacing the needle of the Agilent 1100/1200/1260 Standard Autosampler may be required during regular maintenance, replacing a bent needle, or flow path blockage...
This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosamplers
Issue
Replacing the Needle Assembly of the Autosampler is required when:
The needle is visibly damaged.
Leaks or blockages in the...
This Information Applies To: Agilent 6000 Series LC/MS Systems equipped with MS40+ rotary vane pumps
Issue
Replacing the MS-40+ (rough pump) fluid periodically is a key maintenance procedure to improve vacuum performance and minimize the possibilities...
This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. For instructions for removing an ion capillary on LC/MS instruments with a gate valve installed, please click here
Issue...
This Information Applies To: Agilent 6000 series LC/MS instruments
Issue:
Cleaning the ion transfer capillary as part of regular maintenance or as part of troubleshooting when a loss of sensitivity in the instrument occurs.
Steps to follow ...
This Information Applies To: Agilent 6000 Series LC/MS systems.
Issue
Efficient electrospray ionization (ESI) in your LC/MS ion source requires a properly adjusted nebulizer needle in good condition. The nebulizer needle is a consumable item in...